[time-nuts] HP 5061B - Beam Current
Scott McGrath
scmcgrath at gmail.com
Sun Oct 5 16:45:59 UTC 2008
Did you try the LF injection test?
On Sun, Oct 5, 2008 at 8:48 AM, Jeffrey Okamitsu <w3kl at w3kl.com> wrote:
> I've spent a good deal of time this weekend working with the 5061B I recently acquired.
>
> The unit appears to be making stable 10 MHz. I've compared it the the Rubidium source I have and the two compare as expected.
>
> There are two problems:
>
> 1) Beam current is low - even after going carefully through the alignment procedure outline in the service manual, the best I can get (Loop Open, Mod Off) is a bit over 10 nA.
> 2) The "Continuous Operation" LED will not come on after pressing Logic Reset. I used the trouble shooting guide to diagnose the LED and it's pointing to the circuit on the A14 board. This is as far as I have gotten so far. Note that the trouble shooting guide is suggesting that it might not be a problem with beam current, but an electronics problem (the LED driver).
>
> Anyone have any insight?
>
> Thanks for reading.
>
> Jeff
>
> Jeffrey K. Okamitsu, PhD, MBA
> +1-609-638-5402
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts