[time-nuts] HP 11729C versus 11848A
Adrian
rfnuts at arcor.de
Sat May 8 05:38:29 UTC 2010
Just the roughly 16 dB of insertion loss caused by the 562 ohm resistor
at the input don't make that filter such a great solution ;-)
So, I brewed something better together...
The 10 kHz beat note is now near 0 dBm when the R input signal is
decreased by 40 dB, which makes a lot more sense than before...
And, I have a noise floor of some -165 to -169 dBc/Hz at +7.5 dBm input,
and I'm seeing the 10811A's some 5 dB above that.
Adrian
John Miles schrieb:
>> John,
>>
>> I'd say you nailed it.
>>
>> After some more testing, I can confirm that the limiter amp and the LPF
>> are the culprit.
>> I opened the box and plugged directly into the mixer LO port.
>> And, for the LPF, as a quick 'n dirty solution, I connected the <1 MHz
>> front panel output with the LNA input.
>> Now, at 10 dBm each into the mixer ports, I'm getting a noise floor of
>> <-145 dBc/Hz at 100 Hz and about -170 dBc/Hz at 10 kHz and above.
>>
>
> That's pretty aggressive for 10811s. The floor on those is usually
> around -165 dBc/Hz. How's your calibration process -- are you accounting
> for the 600 ohm output Z of the <1 MHz output port? It'll lose a few dB if
> you try to drive 50 ohms with it, and/or the filter response won't be
> correct.
>
> -- john, KE5FX
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
More information about the time-nuts
mailing list