[time-nuts] HP5065A Rubidium 2nd Harmonic Readings

John Miles jmiles at pop.net
Fri Oct 29 04:34:04 UTC 2010


How's the photo current?  Does it vary as well?  If not, I don't think I'd
be too worried about the physics package.

I would, however, check the electrolytics on the oven controller board.  You
don't want an oven controller failure.

-- john, KE5FX

> -----Original Message-----
> From: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com]On
> Behalf Of Brian Kirby
> Sent: Thursday, October 28, 2010 9:15 PM
> To: precise time
> Subject: [time-nuts] HP5065A Rubidium 2nd Harmonic Readings
>
>
> I recently had a long term power failure and ran out of UPS time.  The
> power was out over a day and when I returned home, it had been restored.
>
> The rubidium recovered, I hit the logic reset button and start and got
> the green light.
>
> I made some meter readings, and noticed something I have not seen
> before.  The 2nd Harmonic reading is showing 30 and its bouncing up to
> 31 about every 2 to 3 seconds.
>
> Otherwise, the unit seems to be performing nominal.
>
> Anybody ever seen this ?  Hope its not trying to tell me failure of the
> cell is on the way...
>
>
> Brian - KD4FM
>