Bob,
Can you re-run your test with a different length of loop-back cable?
Something between a few more feet and a few more inches is fine.
/tvb
>I ran my 5370B for one hour, counting its own reference with a one second
> gate. The summary of the results are:
>
> min 9999999.99650 Hz
> max 9999999.99963 Hz
> avg 9999999.99796 Hz
> median 9999999.99787 Hz
>
> Per the manual's Performance Test 5 (page 4-12) the spec is +/- 0.005 Hz.
> It would be nice if my average was 10,000,000.000 HZ but?? From a Time-nuts
> thread of several months ago, the adjustment for this result is very touchy
> and it's likely you will make the result worse rather then better. Since
> mine is in spec, I'll leave it alone.
>
> I have sent the graphical & raw data to Corby, if anybody else is interested
> please advise me of your email address.
>
> Regards,
> Bob Martinson, N1VQR