[time-nuts] SRS SR620 External Source Issue -- Help Request
Mike Fahmie
mpfahmie at lbl.gov
Wed Feb 21 18:50:59 EST 2007
At 02:57 PM 2/21/2007, you wrote:
>I have an SR620 counter that I set up with a Z3801A as an external
>reference. If I put a bnc T connector at the output of the Z3801A and use
>two equal length bnc cables, one to the ext. ref input on the back and the
>other to channel A then do a frequency measurement, I get a mean that is
>about .0015 Hz below 10,000,000.0000Hz.
>
>Does any know why this might be happening? I would expect it to read
>10,000,000.0000 exactly give or take a couple on the last digit.
Perhaps you're running into some sort of meta-stability problem since your
clock and input are synchronous. Try lengthening one of the cables a foot
or so and see if it makes any difference.
Also, it could be the compound result of noise in both the reference input
and signal input amplifiers.
-Mike-