[time-nuts] New Time nut FRK-H repair questions
Rex
rexa at sonic.net
Thu Nov 22 14:53:16 EST 2007
Corby gave you good information. Probably you need to monitor the
resistance to ground on the lock signal for low (mine was about 140
ohms) when locked. In the manual referenced below, it looks like the
lock output is pulled up to +17V so you *would* see a voltage change if
yours had that circuit. Mine had no voltage unless is was pulled up
externally, just a high resistance unlocked and low when locked.
The 12.2 V tune monitor voltage sounds high and you should probably try
to adjust the xtal trimmer to bring the voltage down, as Corby described.
In case you do have a problem, here is a link to a manual:
http://www.to-way.com/frk.pdf
I have a paper copy of an FRK (H or L) LN manual which matches the one
I have, and the circuits are a bit different, so see how close things
look if you need to get inside. The general operation and theory should
be the same.
The FRK I had did not lock when I got it. The xtal oscillator in mine
had aged to the point where the trimmer could not achieve lock even at
extremes of range. I had to change a fixed capacitor in the trimmer
circuit. In the manual (linked above) that would be C-11 on page A21. On
the FRK I had the circuit was different and it was C-7. I actually had
to remove it completely to get the tuning range centered on mine.
If you decide yours is not locking and if you have access to an accurate
calibrated counter, you should be able to set the tuning voltage to a
fixed center value with a potentiometer soldered in (as described in the
manual) and adjust the trimmer for 5 MHz output. If it was far from 5
MHz to begin with (more than about +-15 Hz) the adjustment should let it
lock.
When I got mine I didn't have a counter I could trust, so I monitored
the servo signals on TP-2 or TP-3 for the 127/254 Hz indication to see
when it was locking.
Hope some of that may help.
-Rex
swingbyte wrote:
> Hi, I've been reading this group for some time now and have recently
> purchased an Efratom FRK-H unit. Now I need to make it work! I have
> powered it up and after warmup it draws ~400mA at 26.5V. I have a 10MHz
> output sinewave and a lamp voltage of 8.14V. The xtal ctrl voltage is
> 12.2V and the lock is 0 v. I haven't been able to find out what the
> lock voltage should be for correct operation, but having the xtal
> control voltage so high makes me think that if might be at its limit.
> Can someone suggest my next course of action?
>
> Thanks for your help
>
> Tim
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
>
More information about the time-nuts