[time-nuts] 35601A as stand-in for 11848A ?
    Bruce Griffiths 
    bruce.griffiths at xtra.co.nz
       
    Sun Jan  6 17:52:41 EST 2008
    
    
  
SAIDJACK at aol.com wrote:
> In a message dated 1/5/2008 16:58:01 Pacific Standard Time, jmiles at pop.net  
> writes:
>
>
>   
>> That's a good point; how are you (Said) calibrating the  system?
>>     
>
>   
>> For simplicity I'd drive the mixer's LO port at its full  spec and start with
>> the RF port unsaturated, at (presumably) 0  dBm.
>>     
> Hi John etc,
>  
> I've been reading all the posts and greatly appreciate the help. I may end  
> up getting a 11729C, these seem to be reasonably priced.
>  
> I've tried two ways to calibrate my setup: one was to detune the OCXO's by  
>   
>> 20Hz, and measure the resulting beat frequency level.
>>     
>
> Another was to use our FireFox signal generator to add a 10.0001MHz  signal 
> -30dB below the carrier into the DUT output using a  power combiner, and 
> measuring the resulting 100Hz spur.
>  
> Both methods result in output spurs that are within <2dB of  expected.
>  
> I will try the hints from everyone,
> thanks,
> Said
>
>   
Said
Another issue to consider is mixer dc offset and mixer induced phase shift.
Both of these effects will cause the PLL to lock with the RF and LO
signals not in exact phase quadrature (unless the correct dc offset is
used at the PLL input amplifier).
The phase relationship between the RF and LO signals at lock is easily
determined with a time interval counter or an oscilloscope.
When selecting a mixer low dc offset is desirable however this is
usually not stated in the datasheet except for those mixers specified
for phase detector operation.
However the dc offset can be estimated from the mixer port to port
isolation specs see:
http://www.wj.com/documents/Tech_Notes_Archived/Mixers_phase_detectors.pdf
There are also other effects such as mixer induced phase shift which can
be more significant at higher mixer input frequencies.
The ideal is to adjust the mixer operating point to minimise its AM
sensitivity, this doesn't necessarily correspond to phase quadrature
between the RF and LO signals at the RF and LO ports.
Saturating both mixer ports can help minimise its AM sensitivity.
Bruce
    
    
More information about the time-nuts
mailing list