Why an 8082A, that's a 250MHz unit. Would not an 8015A 50MHz work. I think
it's the same only with a lower bandwidth.
----- Original Message -----
From: "wje" <wje at quackers.net>
To: "Discussion of precise time and frequency measurement"
<time-nuts at febo.com>
Sent: Thursday, June 19, 2008 9:01 AM
Subject: Re: [time-nuts] 5370B arrived - any FAQ
> Why, you use your 8082A! I got one for a steal a couple of years ago on
> EBay.
> However, all you really need is a fast rise-time pulse generator that
> has external triggering and a variable trigger delay, and can generate
> a +1/-1 volt swing. The frequency the tests use aren't that critical,
> anything 1Mhz or higher should work. The only semi-critical parameter
> is the rise time; the faster it is, the less trigger uncertainty you'll
> have. 1ns is good. :) Also, unless the delay time setting is rather
> fine, it's extremely difficult to actually do the anti-coincidence
> alignment.
> What I'm missing is a spectrum analyzer, so I haven't done the
> frequency multiplier alignment on either of my 5370s. Any tricks for
> doing that without one?
> Bill Ezell
> ----------
> They said 'Windows or better'
> so I used Linux.
>
> Chuck Harris wrote:
>
> wje wrote:
>
>
>
> Freq vs time interval - simplistic answer: when you're using
> time-gating, you're looking at more samples than in averaged one-period
> mode. For 10Mhz, a gate of 0.1 secs is 1 million periods. In period
> mode, you're averaging a maximum of 100,000 periods. Accuracy is
> proportional to the square root of the number of samples.
>
>
> Also, the trigger point jitter when in one-period mode affects every
> sample, but when in gate mode it only affects the first and last
> cycle.
>
>
> The 'specifications' section of the manual gives the formulas for
> determining both resolution and accuracy (which are NOT the same) for
> the various modes.
>
> I would suggest you go through the detailed alignment procedures even
> though the unit passes the operational checks. These units almost always
> have drifted out of alignment in the analog front-end unless you were
> lucky enough to get a freshly-calibrated one. You can usually
> significantly improve channel-to-channel trigger consistency and lower
> the jitter by doing so.
>
>
> What does one do where the tests call for an HP8082A pulse generator?
> The silly things are so rare that they are priced in the stratosphere!
>
> -Chuck Harris
>
> _______________________________________________
> time-nuts mailing list -- [1]time-nuts at febo.com
> To unsubscribe, go to
> [2]https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
> References
>
> 1. mailto:time-nuts at febo.com
> 2. https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the time-nuts
mailing list