[time-nuts] 5061A troubleshooting
Scott McGrath
scmcgrath at gmail.com
Tue Sep 30 14:14:54 EDT 2008
How does one increase the EM voltage?
I do have the manual and I have done the LF test but the ionizer and
mass spec voltages were what I was looking for
On Tue, Sep 30, 2008 at 1:11 PM, corby d dawson <cdelect at juno.com> wrote:
> Scott,
>
> The -.210 v one the multiplier is OK for now. The tube value is for when
> the tube was originally tested. As the electron multiplier in the tube
> ages the required voltage needs to be increased (up to a maximum of
> 2500VDC).
>
> If you have a manual I would suggest that you do the low frequency test.
>
> It takes a frequency at 1/2 the zeeman frequency that you apply to a pin
> on the underside of the unit on the long terminal board.
>
> This allows you to test the tube even if the RF circuitry is at fault.
>
> If the low frequency test fails then before blaming the tube you need to
> get in and measure the ionizer voltage (older units a 1VRMS square wave,
> new units 1VDC) and also measure the mass spectrometer voltage with the
> pot at max CW (about +17 to 18VDC.
>
> If everything checks out but low freq. test still fails increase the EM
> voltage to 2500VDC and repeat the test.
>
> If no low frequency response by now the tube has a high probability if
> beaing dead!
>
> If the test is good then your problem lies in the RF circuitry
> somewhere.
>
> Corby Dawson
> ____________________________________________________________
> Click here to find affordable degree programs.
> http://thirdpartyoffers.juno.com/TGL2141/fc/Ioyw6i3l8Op2dKMrmSXhHChrYqlJXFP6LXoDtx9dtYTJzdmc3WCvct/
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
More information about the time-nuts
mailing list