[time-nuts] HP E2050A LAN/HPIB Gateway
Adrian
rfnuts at arcor.de
Tue Sep 1 22:43:01 UTC 2009
Matt,
I have the E5810A manual (that can be downloaded from Agilent).
Thought it wouldn't work, but...
Finally, I found the E2050A default IP address 192.0.0.192 somewhere on
the web.
Then I connected the E2050A and my PC directly, using a crossover cable.
Some important instructions are on p. 49 of the E5810A manual...
Enter in a DOS window: route add <192.0.0.192> <PC IP-Address>
From that point I got access to the configuration utility via Telnet
(must be activated, btw.).
I assigned a fixed IP address etc. on my router and entered it into the
E2050A.
Exit the Telnet utility with 'reboot' to save the changes.
Connect the E2050A with a non crossover cable to the local network,
cycle power, and it works!
As usual, the biggest headache was to get the basic communication
established...
I'd still be interested in the original manual, but I'm happy it works
without.
Adrian
Matt Ettus schrieb:
> I'm not sure if the programming is the same, but we use a couple of
> the E5810A and love it. We have full python code for using it if that
> is of use.
>
> Matt
>
> On Tue, Sep 1, 2009 at 10:32 AM, Adrian<rfnuts at arcor.de> wrote:
>
>> Ulrich,
>>
>> thanks for your reply.
>>
>> I have installed the I/O libraries suite 14.0 that supports the E2050A.
>> However, I don't get any communication with the unit, no matter if I connet
>> it to my LAN or directly with a crossover cable. I think a manual would
>> probably do the trick.
>>
>> Adrian
>>
>> Ulrich Bangert schrieb:
>>
>>> Adrian,
>>>
>>> not sure whether you can reach your aim completely without a manual.
>>> However
>>> installing the AGILENT I/O Libraries should bring you a big step forward.
>>> Having installed the software it should recognize the E5020A and the GPIB
>>> devices connected to it, so that some basic communication tests should be
>>> possible from within the GUI (i.e. without any programming).
>>>
>>> After basic functionality tests you should be able to talk to your
>>> instruments through the I/O Libraries with any software that knows how to
>>> make use of VISA functionality, for example my EZGPIB utility (but of
>>> course
>>> there are others too).
>>>
>>> Best regards
>>> Ulrich
>>>
>>>
>>>
>>>
>>>
>>>> -----Ursprungliche Nachricht-----
>>>> Von: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com] Im
>>>> Auftrag von Adrian
>>>> Gesendet: Dienstag, 1. September 2009 15:48
>>>> An: time-nuts at febo.com
>>>> Betreff: [time-nuts] HP E2050A LAN/HPIB Gateway
>>>>
>>>>
>>>> Hi,
>>>>
>>>> hope it's not considered too far off-topic. (Actually, I'm trying to
>>>> control some counters etc. of my time and frequncy rack through the E2050A.)
>>>> I just don't get it working. I have the 4-page 'overview' but couldn't find
>>>> any other information.
>>>>
>>>> Anyone got a manual for me?
>>>>
>>>> Regards,
>>>> Adrian
>>>>
>>>> _______________________________________________
>>>> time-nuts mailing list -- time-nuts at febo.com
>>>> To unsubscribe, go to
>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>> and follow the instructions there.
>>>>
>>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts at febo.com
>>> To unsubscribe, go to
>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>>
>>>
>>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
More information about the time-nuts
mailing list