[time-nuts] reference oscillator input circuit
Rick Karlquist
richard at karlquist.com
Wed Dec 8 20:13:06 UTC 2010
Yes, and thus if you are going to use a comparator or line
receiver (not recommended), then you should use the SLOWEST
one that still works, if you want to optimize jitter.
This is because the noise bandwidth is less.
Bruce Griffiths wrote:
> One can estimate the resultant jitter from the input slew rate and the
> circuit noise and bandwidth.
> Too much bandwidth with a slow input slew rate increases the jitter
> substantially over that possible with an optimal circuit.
>
> Bruce
>
> Robert LaJeunesse wrote:
>> Assuming a transformer coupled input (with biasing via a secondary
>> center tap)
>> why not use a fast differential PECL to CMOS level translator? For
>> example, the
>> IDT ICS508 will take 0.3 to 1.0 V p-p input and give 2.5, 3.3, or 5 V
>> swing on
>> the output. The chip works down to DC and keeps the duty cycle in the
>> 40%-60%
>> window up to 250MHz (at 3.3V out). Jitter and noise is not spec'd
>> however.
>>
>>
>> To increase the noise immunity with a relatively slow 10MHz sine source
>> I'd look
>> at boosting the amplitude with the transformer, then clipping with
>> balanced
>> series resistors and back-to-back diodes so the translator sees a higher
>> dV/dT
>> on its inputs.
>>
>> Might want to look in some old Motorola ECL appnotes for other possible
>> schemes.
>>
>> Bob
>>
>>
>>
>>
>> ________________________________
>> From: jimlux<jimlux at earthlink.net>
>> To: Discussion of precise time and frequency
>> measurement<time-nuts at febo.com>
>> Sent: Wed, December 8, 2010 10:31:08 AM
>> Subject: [time-nuts] reference oscillator input circuit
>>
>> I'm looking for suggestions on a general circuit that can be used to
>> receive an
>> external frequency reference (nominally a real clean sine wave at, say,
>> 10 MHz,
>> although up to 100 MHz is possible) and turn it into a "real clean"
>> square
>> wave. Galvanic isolation is a plus (a transformer or capacitor would
>> probably
>> do that).
>>
>> I was thinking about rummaging through the schematics for test equipment
>> reference inputs (since they've already "solved" the problem, eh?), but
>> any
>> other ideas would be welcome.
>>
>> I've scanned the archives of time-nuts, and while we have a fair amount
>> of
>> discussion on how to square up the 1Hz (or 100Hz) in a phase noise/ADEV
>> setup,
>> not so much on what to do with the 10 MHz. Rick has commented that you
>> don't
>> want to use a comparator. I have the papers by Dick, et al, and Collins,
>> as well
>> as all the others.. they tend to be looking at the low frequency
>> problem,
>> although the analysis is certainly applicable.
>>
>> I don't know that I'm looking for the whole multiple limiting stages
>> scheme in
>> any case.
>>
>> Oh, as far as performance.. Say the need is to not horribly degrade a
>> good
>> quality crystal oscillator... here's a typical set of specs:
>> 76 MHz
>> 1Hz<-90dBc
>> 10Hz<-110dBc
>> 100Hz<-120dBc
>> 1k-100k<-125dBc
>>
>> Adevs of the oscillator run from 5E-12 at 0.1 sec, down to 1E-12 at 10
>> sec, and
>> back up to 2 E-12 at 1000sec.
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>>
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
More information about the time-nuts
mailing list