Hi

Definitely time to scrap the counter. To save you the excessive landfill
charge I'll take care of the disposal for you. 

Unfortunately for my future disposal business, the answer to your question
is indeed, yes you can likely fix it yourself. The answer to the other
question is also yes, this is pretty typical of a 620 after auto cal. 

There are a bunch of things in the archives about which registers to do what
to in order to make the SR620 "run right". Some of them are fairly tedious
back and forth tweaks. As I recall there are at least two approaches to
taking care of the problem, depending on how far you want to go. 

Bob

-----Original Message-----
From: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com] On
Behalf Of SAIDJACK at aol.com
Sent: Wednesday, February 23, 2011 5:41 PM
To: time-nuts at febo.com
Subject: [time-nuts] SR-620 cal

Hi guys,
 
need some help with a cal on my SR-620 counter.
 
In frequency measurement mode the unit always has an offset of up to  
+/-2E-010 after auto-cal when feeding the same 10MHz to the (A) input and to
the  
external reference. It should not have any offset of course.
 
Stanford says "the accuracy spec is +/-3.5E-010, so your unit is within  
spec". Sigh. that's a crappy spec I should say..
 
Is this normal? On the Agilent counters I get 0.0 perfectly when  averaged 
when doing that test with the carrier subtracted, no matter what gate  time 
I use.
 
Any help would be great, as I have to decide if I pay Stanford to do the  
cal, and they won't guarantee that the unit will perform any better 
whatsoever  after spending $500 for the calibration..
 
Thanks,
Said
 
 