[time-nuts] Antique Rubidium Standard Questions
Magnus Danielson
magnus at rubidium.dyndns.org
Mon Apr 23 20:08:08 UTC 2012
Ed,
On 04/23/2012 07:54 AM, Ed Palmer wrote:
> I'm playing with a Tracor 304-B Rubidium Standard from 1969. I'm using
> it as a learning exercise to find out more about the guts of a Rubidium
> standard and how it works.
>
> This thing is a beast! Rack-mount, 3U high, 39 pounds (~18 kilos), 9
> plug in circuit boards. The OCXO is an oddball frequency that is
> multiplied directly to 6.8 GHz. There's no synthesizer in that chain. A
> synthesizer is used to convert the oddball frequency to a 5 MHz output.
OK. A bit different to some of the rubidiums, but looking around it is
not as uncommon as one might think. This technique is used in some of
the FEI 5680 and also the later GPS rubidiums.
> It's sort of working. The error signal isn't up to spec, but it's strong
> enough to give a stable lock although there's no trace of a second
> harmonic signal. Allan Deviation is in the Xe-12 range from 1K to 10K
> seconds. The OCXO has a not-yet-resolved issue that is probably
> degrading the results.
What is the OCXO issue?
> The lock frequency suggests that the Rubidium cell has drifted down by
> ~30ppt over the 40+ years since it was built. Is that reasonable? That's
> much more drift than the specification states, but I doubt if the spec
> was intended to be valid for 40 years!
>
> Could the drift be at least partially responsible for the lack of second
> harmonic? A message on the list (
> http://www.febo.com/pipermail/time-nuts/2006-April/020562.html ) said
> that you could peak the second harmonic by adjusting the cavity tuning.
> If the cell and the cavity are out of sync would that kill the second
> harmonic? How close to they have to be? If this thing has a cavity
> tuning adjustment I haven't found it.
You should see second harmonics regardless, it's an effect of the dip
itself. I would check if you can observe the second harmonics on the
signal from the detector by some other means. If you seems to lock up
but does not see second harmonics, it just could be something in the
second harmonics detection which needs some LTC.
There are many sources for shifting the frequency, including the power
supply (check voltage and clean-ness), the trimming of the C-field, the
leakage of helium, resonator tuning, temperature trimming of lamp and
resonator may shift amplitude and hence frequency through light pulling.
It would be good if you could hook up the modulation sine on X and
return signal on Y on a scope, that would give you a clear display of
the resonance dip.
Oh, and check if you have leakage problems around the integrator cap,
that would also shift the frequency.
Cheers,
Magnus
More information about the time-nuts
mailing list