[time-nuts] EIP545A 18GHz counter query
shalimr9 at gmail.com
shalimr9 at gmail.com
Sun Dec 9 12:22:34 UTC 2012
Good practice with a spectrum analyzer is to put a 20 dB pad at the input all the time unless you are sure you don't need it for a particular measurement.
It has saved my bacon many times.
It also protects the input connector from excessive wear.
Didier KO4BB
Sent from my Droid Razr 4G LTE wireless tracker.
-----Original Message-----
From: Chris Wilson <chris at chriswilson.tv>
To: Chris Wilson <chris at chriswilson.tv>, Discussion of precise time and
frequency measurement <time-nuts at febo.com>
Sent: Sun, 09 Dec 2012 5:54 AM
Subject: Re: [time-nuts] EIP545A 18GHz counter query
> Did do some more reading and I think what I suggest is the right approach.
> Band 2 and 3 use the same VCO. So if band 2 is generally working
> then its possibly the rf multiplier yig filter mixer diodes etc. But
> if the RF detectors show RF then its going to be a much more interesting issue.
> I do not know what sort of test equipment you might have. I have
> spectrum analyzers and such. And do agree without extenders that I don't have its a royal pain.
> Regards
> Paul.
09/12/2012 11:45
Thanks very much for your input with this, much appreciated. I do have
a spectrum analyser, but it's a bit like having a set of surgeons
implements. Interesting to have, but needing the requisite skill to
use safely :) I can do only the most basic stuff, to be frank I am
very wary of blowing the input stage up by doing something daft!
I think I follow your train of thought here Paul. But, before seeing
your post I did some tests at TP4, to see if the unit was constantly
scanning for a signal when on Band 2. here is what I found and saw:
I have now refitted the new Q12, and removed the test 1k resistor.
OK, the trace from TP4 does different things depending what band it is
on. Bands 1 and 3 it just sits there, #1. On Band 2 it gives what appears
to be two traces, #2. Changing the timebase gives the other trace I show
linked, #3. On bands 1 and 3 only a single trace appears. So Band 2 is
doing something different at TP4, with no input signal. Warm Q12 with
an input signal to Band 2 and the two traces change to a single trace,
#4.
Once it cools again the two traces return.
Traces:
#1 http://www.gatesgarth.com/scope/TP4-band1.jpg
#2 http://www.gatesgarth.com/scope/TP4-band2.jpg
#3 http://www.gatesgarth.com/scope/TP4-...t-timebase.jpg
#4 http://www.gatesgarth.com/scope/TP4-...put-signal.jpg
I would need explicit instruction to poke about using the SA. It's an
HP 8568B. Thanks again Paul :)
--