[time-nuts] DMTI and Timelab setup
John Miles
jmiles at pop.net
Thu Dec 13 04:21:50 UTC 2012
Hi, Luciano --
The sampling interval would be whatever rate the counter is returning
readings, as usual. I usually let the acquisition driver estimate the rate,
unless I know it's exactly one reading per second.
In frequency mode, the scale factor should be 1E-6. You need the program
to treat the frequency readings as microhertz, to force the 1000000:1
heterodyne factor to be accounted for when the frequency readings are
integrated back to phase differences.
For TI readings the scale factor should also be 1E-6, and you would enter
5E6 as the input frequency. I *think* that'll make it handle phase wraps
properly, but you might send me a few plots to make sure. I've never tried
a DMTD with TimeLab before.
-- john
> -----Original Message-----
> From: time-nuts-bounces at febo.com [mailto:time-nuts-
> bounces at febo.com] On Behalf Of Timeok
> Sent: Wednesday, December 12, 2012 5:15 AM
> To: time-nuts at febo.com
> Subject: [time-nuts] DMTI and Timelab setup
>
> Hi all,
>
> if I will use a dual mixer TI meter, suppose to have :
>
> LO 5.000.005 Hz
> Ref and Ftest 5 MHz
> IF= 5Hz
>
> how do I have to setup Timelab for acquisition?
>
> Sampling interval = ....
> Input frequency = ....
> Scale factor = ....
>
> Thanks all,
> Luciano
> timeok
>
> _______________________________________________
> > time-nuts mailing list -- time-nuts at febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> > and follow the instructions there.
>
> --
> timeok
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-
> bin/mailman/listinfo/time-nuts
> and follow the instructions there.