Ok, thanks for clarifying. In general the time constant one chooses must reflect both the intrinsic performance of the OCXO (essentially constant) and the realities of GPSDO mechanical, sky-view, and environmental conditions (possibly variable). Disabling an oven during a run is equivalent to a radical change in environment and not re-tuning the loop parameters will lead to sub-optimal or misleading results when plotted.
If you have time, it would be instructive to re-run the experiment. First with double oven enabled and do your best case ws-tuning. Then disable the outer oven and again do a best-case tuning. The phase/freq/adev plots would be revealing, as well as the (major?) difference in optimal tuning values.
/tvb (iPhone4)
On Jul 14, 2013, at 9:19 PM, "WarrenS" <warrensjmail-one at yahoo.com> wrote:
> Tom
> 
> My posting and plot was only meant to show the difference in tempcoef between an undisciplined single and dual oven 10811 osc which in this case is clearly =>  60 to 1.
> Your comments  bring up a different subject which is who needs it and how good does a controlled GPSDO oscillator need to be when not in holdover.
> 
> As you know, the purpose of a GPSDO control loop is to make the oscillator's long term stability relatively un-important.
> The longer the measurement time the less important the stability of the controlled osc is in a GPSDO, and as time increases past the GPSDO control loop time constant, the osc stability matters less and less
> 
> What you are seeing and saying when analyzing the phase and Freq errors plots, is closed loop performance.
> The phase and freq plots of the dual oven osc would pretty look the same even if compared with a 'perfect' osc, because the dual osc plots is already near or at the noise floor of that TBolt setup and antenna.
> 
> One can measure the longer term stability of an oscillator different ways;
> 1) Hold the EFC voltage constant and measure the change in frequency or phase with time.
> 2) Measure the scaled EFC change necessary to hold the oscillator's freq or phase output constant
> When done carefully and with the EFC voltage scaled correctly both ways can give the same answer.
> 
> Answer1)
> The way I measured the two tempco's is by measuring the correlation between the EFC control voltage and the temperature plot
> In the case of the single oven osc, the plot gains are set so that when overlayed the EFC DAC plot looks as close as possible to the temperature plot.
> When the plot time is >24 hr and there is good repeatability, the TC is just the ratio of the two plot gains, i.e theEffective EFC freq change divided by the delta temp.
> In the single oven case DAC plot gain = 1e-10 per division,  temp plot gain = 1.5C per division. Tempco = 1e-10 / 1.5  ==  6.7 e-11 / degC.
> I did the same thing for the dual oven trace by expanding the gain and zero 
    
    
More information about the time-nuts