[time-nuts] HP Agilent 34401A

Volker Esper ailer2 at t-online.de
Sat May 18 07:50:30 EDT 2013


Thank you, Bob, yes, that was my first guess, too. I opened it and 
shortened the relay contacts with an extremly low ohms short and the 
display changed significantly when coming to K101. I have to get one.

Besides, after fumbling and knocking and all that handling the 
instrument is now quite stable - that lasts till the next hyper 
important unattended long time measurement...

Thanks a lot

Volker


Am 18.05.2013 03:46, schrieb Bob Camp:
> Hi
>
> Never seen that problem, but it sounds very much like a relay / mux problem in the ohms section.
>
> Bob
>
> On May 17, 2013, at 9:23 PM, Volker Esper<ailer2 at t-online.de>  wrote:
>
>    
>> Hi,
>>
>> I bought a 34401A lately. It's all fine, except it's got a fault when measuring low ohm with 2-wires: the value runs (monotonic) from 0 to 30 oder 40 Ohms at a speed of say 1 ohm per second and after a while it decides to run back... When using 4-wires-measurement everything's alright.
>>
>> Did anyone have that problem before?
>>
>> Thank you
>>
>> Volker
>>