[time-nuts] comparing two clocks
SAIDJACK at aol.com
SAIDJACK at aol.com
Sat Feb 22 20:49:50 EST 2014
Jim,
when I did the test on the 53132A, I did the test with the two signals on
top of each other with a very small cable offset of 400ps, then I added a
10ns delay line to the B signal just to see if the counter would behave
differently. Here are the results, pretty much looks identical with 0.4ns offset
or 10ns offset.
bye,
Said
In a message dated 2/22/2014 10:06:30 Pacific Standard Time,
ailer2 at t-online.de writes:
Jim,
If I get you right, you want to compare the 10MHz outputs (not the
1PPS). As Jim and Bob told us so far, the thing is to provide, that
input A _always_ starts before input B (or the other way around).
Connect the signals to an oscilloscope, and check, how much the phase
differs - if the rising slopes occur close together, put some
meters/yards of coaxial cable into one of the two signal paths. 1 meter
is roughly worth 5ns - while the period of 10MHz is 100ns, 1m cable will
phase shift about 18 degrees. I didn't verify, if the coax cable (with
it's microphonic effect) affects the ADEV - does anybody have experience
with this? Otherwise I'd have to fire up my counter and have a
measurement on the run...
Of course, inverting one signal will do as well. If you do it with extra
electronics that definitely will affect the ADEV. I find it much easier
to use some meters of cable.
Ok, my counter is heating up by now...
Volker
Am 22.02.2014 14:17, schrieb Jimmy Burrell:
> I need some help with a 'noob' question regarding some practical
examples in some of the NIST literature. When attempting to compare two clocks,
I'm a bit confused on the subject of exactly how to use my counter to compare
a delayed clock relative to another. Or perhaps I should just say
'comparing two clocks'. Let's take some concrete examples.
>
> Let's say I want to characterize my Morion MV89 ocxo using my HP5335a.
Obviously, I can tune the MV89's 10MHz by +/- 1Hz and feed it to the
counter's input 'A'. Obviously, I can feed in a second, external reference clock
at 10MHz into input 'B'. Suppose, however, I didn't have an external
reference clock. Can I compare against the counter's internal time base by
hooking a line from the rear jack time base output to channel 'B' input? Or am I
making it too complicated? Do I simply plug into input 'A' and go?
>
> In a somewhat related question, in this article
(http://www.wriley.com/Examples%20of%201%20PPS%20Clock%20Measuring%20Systems.pdf) where two
clocks, both divided to 1PPS, were compared, W.Riley makes the following
statement, "The two 1 PPS outputs were connected to a Racal Dana 1992 time internal
counter having 1 nanosecond resolution, and the start and stop signals
were separated sufficiently in time for the counter to function properly". I
wonder what exactly is meant by "separated sufficiently in time for the
counter to function properly" and how one would go about doing this? For
example, is inverting one of the signals sufficient separation? If not, how is
this typically done? Delay line?
>
> Thank you,
>
> Jim...
> N5SPE
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
_______________________________________________
time-nuts mailing list -- time-nuts at febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
-------------- next part --------------
A non-text attachment was scrubbed...
Name: 1PPS_jitter.gif
Type: image/gif
Size: 16473 bytes
Desc: not available
URL: <http://www.febo.com/pipermail/time-nuts/attachments/20140222/5ebaab28/attachment-0001.gif>
More information about the time-nuts
mailing list