[time-nuts] HP 8568B

Tommy phone tholmes at woh.rr.com
Sat Jun 28 22:24:22 EDT 2014


Hi Hui...
The first thought is that you are tied to the resolution of the span divided by the number of displayed points -1. Try a narrower span. If you look at the absolute accuracy spec that error may be within the '66's capability.

Also, I believe there is a marker counter function that will give you a more accurate reading.
Another consideration is that not all of the internal LO's are locked to the 10 MHz reference.
Don't panic yet :)


From Tom Holmes


> On Jun 28, 2014, at 9:48 PM, "Hui Zhang" <ba6it at 163.com> wrote:
> 
> Hello all:
>    Rencently I bought a second-hand HP8568B Spectrum Analyzer, it kind of vintage but works well. Only problem is its time base lost accurate. When I use it measured my HP Z3801A output (locked to GPS), the center frequency read is 9.999912MHz, about 88Hz diffrence. So I decide to use  my Z3801A for its external reference. When I connected the cable and set the reference switch in rear paneal to EXT, the CRT displayed "EXT REF", I believed it worked. But when I use it measure a nother GPSDO (Trimble ThunderBolt locked to GPS), it still have 42Hz diffrence. I am very sure both of my two GPSDO is good, if I use my HP53132A counter to compare them they will has less 1E-10 (1mHz) diffrence. And then I tie a BNC three-way connect from my Z3801A, one way to 8568B's EXT stand input, one way to the signal input in front panel, the  diffrence is still 42Hz - it's 10.000042MHz. I was fully confused, do I need a calibration after I used external 10Mhz stand? I read the manual again but not fou
> nd any infomation about how to do it. What do I do now? Anyone give me a suggestion? Any infomation will be appreciate.
> 
> 
> Hui Zhang
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.


More information about the time-nuts mailing list