[time-nuts] My NTGS50AA failed
EB4APL
eb4apl at cembreros.jazztel.es
Tue Nov 11 21:06:33 EST 2014
Hi,
Removing the oscillator for testing and replacing it with other if it
was the culprit was my first option. I have a spare Trimble oscillator
that probably came from other NTGS50AA since it still have the foam band
attached, but this oscillator is really aged, it needs 7.91 V to bring
it on spot and the maximum control voltage of the NTGS50AA is 5 V.
I was trying to avoid removing the oscillator but probably it must be
done to clarify things.
Thank you,
Ignacio EB4APL
.
El 12/11/2014 a las 2:40, Mark Sims escribió:
> I have seen this caused by the oscillator not responding to the EFC signal. Fixed it by swapping in a MV-89 oscillator.
> The oscillators used in these units don't output an oven temperature monitor signal.
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts