[time-nuts] Lucent KS-24361, HP/Symmetricom Z3809A, Z3810A, Z3811A, Z3812...
Tom Miller
tmiller11147 at verizon.net
Mon Nov 17 15:01:50 EST 2014
That first spike falls right at 60 Hz. I wonder if your test setup is
picking up some hum?
Tom
----- Original Message -----
From: "S. Jackson via time-nuts" <time-nuts at febo.com>
To: <kb8tq at n1k.org>; <time-nuts at febo.com>
Sent: Monday, November 17, 2014 2:26 PM
Subject: Re: [time-nuts] Lucent KS-24361, HP/Symmetricom Z3809A,
Z3810A,Z3811A, Z3812...
Hi Bob,
yes, the 10MHz plot is rotten, no doubt. The 15MHz plot is quite good till
about 40Hz offset, then it becomes pretty rotten too.
Here is one of my 58503A units (using the 10811 OCXO) as a comparison..
measured against our DROR-IIA (this plot was actually done to show the
DROR-IIA PN, but since that unit actually has less noise and spurs than the
58503A we can simply use it as the reference for this purpose).
The good news is that getting the close-in phase noise to be good is very
hard to do and the unit delivers that out-of-the box already. Filtering out
the noise and spurs above 40Hz offset is pretty easy to do. It should be
fairly straight forward to cobble up a small PN filter for those units to
get
rid of the noise and spurs above 40Hz offset.
bye,
Said
In a message dated 11/17/2014 09:31:46 Pacific Standard Time, kb8tq at n1k.org
writes:
Hi
Here’s the phase noise on the 15 MHz. There are a few spurs, and an very
real hump out at the likely frequency of the Lucent switcher. The 15 MHz
is
pretty clean compared to most /all of the other units I’ve seen on the
surplus market.
I would not multiply this up to 40 GHz with a broadband multiplier. I
would be quite happy to run it into a PLL with a rational bandwidth. You
will
beat the noise on the output with a fairly simple VHF VCXO past 100 Hz. No
reason to have a bandwidth outside the 20 to 80 Hz range.
Math:
15 MHz to 150 MHz -> 20 log (N) -> 20 db.
-140 dbc / Hz shown below at 100 Hz offset -> -120 dbc/Hz
You can get numbers better than -120 dbc/Hz at 100 Hz offset out of a
number of pretty simple VHF VCXO circuits. Bert has one that seems to work
fine
for him.
Bob
--------------------------------------------------------------------------------
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
More information about the time-nuts
mailing list