[time-nuts] Changing ADEV, (was Phase, One edge or two?)
Bob Camp
kb8tq at n1k.org
Sat Oct 25 08:02:40 EDT 2014
> On Oct 24, 2014, at 6:25 PM, Magnus Danielson <magnus at rubidium.dyndns.org> wrote:
>
> Tom,
>
> On 10/24/2014 11:31 PM, Tom Van Baak wrote:
>>>> ADEV most certainly does change with time, even for short tau's.
>>>
>>> Can you elaborate?
>>> Such as when, why, what kind of change, how much change,
>>> at how short of tau's, over how long of time,
>>> and using what type Oscillators?
>>> Do you know what in the freq or Phase plot is causing the ADEV to change?
>>
>> I'm happy to let Bob answer his own claim here. I'm curious as well. Unless he's talking about thermal noise, in which case I now believe him 100%.
>>
>> OTOH, for time intervals of minutes to hours or days, the plotted ADEV can often vary. When in doubt, enable error bars in your ADEV calculations or use DAVAR in Stable32, or use "Trace History" of TmeLab to expose how little or much the computed ADEV depends on tau and N.
>>
>> In general, never do an ADEV calculation without visually checking the phase or frequency time series first.
>
> You should make sure that you remove all forms of systematic effects before turning the residue random noise over to ADEV.
>
> If you have random noise being modulated in amplitude, you need to measure long enough for the averaging end not to have a great impact on the result.
Is days long enough for a 1 second tau? If you define 1,000 x tau as “long enough” you are being way more conservative than just about anybody out there. My claim is that rather than telling everybody to run for 10,000 or 100,000 x tau, simply accept that ADEV does / may change.
Removing this or that *before* you do ADEV can get you on a very slippery slope indeed. Removing drift (either time or frequency) - fine. Removing this or that couple of minutes of data because it makes the result look better, that’s likely to get you in trouble. Your customer (or system, or test setup) isn’t likely to accept a “ADEV is ok most of the time” specification.
Is ADEV a good way to measure temperature stability - no of course not. We do indeed have rooms that vary in temperature. They do impact ADEV on a Rb. Removing the delta temperature related data from your ADEV input is not at all easy (1,000 to 10,000 second data …) and I believe it would mis-represent the part. Running in the real world, it’s going to have that ADEV hump.
Yes indeed you can find FCS papers with all sorts of interesting “adjustments” or no processing at all. The consensus seems to be that if you go past drift correction, you really should have a footnote.
Bob
>
>>> Of the many OCXO type Oscillators that I've tested (HP10811 & MV89),
>>> seldom have I seen any significant change (say greater than 10%),
>>> in the short tau (0.01 sec to 1 sec) ADEV values, after the systematic
>>> type errors are removed. (even when starting soon after turn on)
>>
>> This is not my experience at all. Let's figure out what's happening to you.
>>
>> If all your standards look sort the same from tau 0.01 to tau 0.1 to tau 1 then either you need more oscillators to play with or maybe you have a measurement problem. This is especially true if you are doing post-comparator averaging. Averaging, by definition, tends to remove noise, to smooth things out. If your goal is to measure noise, the last thing you want to do is create any electronics or use any analog or digital or numerical filtering that removes or reduces the very thing you're trying to measure.
>>
>> I remind you of this page http://leapsecond.com/pages/adev-avg/ of the perils of averaging data.
>>
>> For most of the world, there's signal and noise. Signal good. Noise bad. But for us, measuring precision clocks, the noise is the signal. So don't do anything that removes or reduces noise.
>
> Systematic signals is however disturbances for the ADEV.
>
>>> ADEV is used to measure random types of noise so there are of
>>> course the statistical uncertainty variations that are a function of
>>> the number of valid data points. I find that using a minimum of
>>> a thousand points at each tau gives good consistent results.
>>
>> Are you crazy? The minimum is just 3 or 4 or 5 data points. Not 1000! You should not see much difference at 10 or 100 or 1000 points. If so, something is wrong with your measurement model. If ADEV(tau) is *that* dependent on tau, check the frequency time-series. Consider removing drift or using HDEV instead of ADEV. We need to talk. If your logic was true, we'd all have to wait 3 years before we could compute the ADEV of a GPSDO at tau 1 day.
>
> No, he is not crazy on this point. While the algorithm only needs 3 points to produce a value, that value will have so bad degrees of freedom that the confidence interval is WAAAY out there.
> The reason that we don't need to wait 3 years for tau of 1 day is that we learned to use interleaving spans of time. We have since had much more development in the algorithms to improve the degrees of freedom for the same N samples, all in an effort to achieve as small confidence interval as possible. Also, the degrees of freedom achieved varies with the tau0 multiple m, and with dominant noise-type.
>
> A great way to illustrate the point of degrees of freedom and the number of sample-points needed to get tight confidence intervals is to see how the high-tau end of a curve updates in TimeLab and behaves as the jiggeling end of a long rope, and as more samples comes in, the jiggeling end moves towards higher taus, but for a particular tau, the amplitude of the jiggeling decreases until it almost stops. This is the effect of the confidence intervals becoming tighter, the range within the real value is becomes smaller and eventually is very tight.
>
> The modern algorithms like TOTAL and Theo can cram out really impressive degrees of freedom for a particular set of N and m compared to older algorithms, which effectively translates into tighter confidence intervals for the same N and m.
>
> Cheers,
> Magnus
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the time-nuts