[time-nuts] Precise Time transfer and relative position over a short baseline
    Michael Wouters 
    michaeljwouters at gmail.com
       
    Mon Apr 11 04:35:41 EDT 2016
    
    
  
Dear Bruce
Posted too hastily ...
Since you don't care about absolute time, maybe using identical receivers
you might be able to get a few hundred ps synchronization using a PPP
solution,  for example.
Attached is 6 months of data for two Javad receivers, with daily PPP
solutions. The antennas for these receivers are separated by about 400 m.
Cheers
Michael
On Mon, Apr 11, 2016 at 3:00 PM, Bruce Griffiths <bruce.griffiths at xtra.co.nz
> wrote:
> There is a proposal to use multiple light bucket style optical telescopes
> to do Intensity stellar Interferometry over short baselines (up to perhaps
> 1km  or so) by using independent clocks to time tag photon  arrivals. store
> the time tags and process the data off line. Depending on the time tag
> resolution there is a need to measure the time differences between the
> independent clocks to an accuracy in the 1ns to 100ps range. Is there a
> better way of doing this other than using geodetic grade GPS receivers
> capable of GPS carrier phase measurements?Since the local clock flywheel
> oscillators will need to not deviate by more than 100ps or so over the
> several minutes required to perform the carrier phase averaging what sort
> of clock will be suitable apart from a good rubidium standard with a
> cleanup oscillator?
> NB Running fibres or coax between the telescopes isnt an option.
>
> The relative positions of the telescopes has to be known to within a cm or
> so for this to work.
> Bruce
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: PPP.jpg
Type: image/jpeg
Size: 20866 bytes
Desc: not available
URL: <http://www.febo.com/pipermail/time-nuts/attachments/20160411/e1c13b26/attachment-0001.jpg>
    
    
More information about the time-nuts
mailing list