[time-nuts] Timelab, two SR620s and losing samples
    Poul-Henning Kamp 
    phk at phk.freebsd.dk
       
    Sun Jan 17 06:52:13 EST 2016
    
    
  
--------
In message <569B61CF.3030708 at rubidium.dyndns.org>, Magnus Danielson writes:
>> This is a common misunderstanding:  Talk-only does *not* protecting you
>> against timing issues on GPIB.
>>
>> On RS-232, yes, but not on GPIB.
>
>Agree, to some degree. It's not a guarantee.
>
>I think you should develop that line of thought, to detail why it helps 
>on GPIB and why not on serial.
It's really very simple:  RS-232 sends blind, you don't even need to
know if there is a receiver or what it does.  If the receiver cannot
keep op, data is simply lost.
GPIB handshakes every byte, so the actions of the receiving end affects
the transmitting end - in particular if the receiver cannot keep up.
-- 
Poul-Henning Kamp       | UNIX since Zilog Zeus 3.20
phk at FreeBSD.ORG         | TCP/IP since RFC 956
FreeBSD committer       | BSD since 4.3-tahoe    
Never attribute to malice what can adequately be explained by incompetence.
    
    