That's easy, Magnus. Do not use a Fluke counter :-)
Don

On 2016-10-09 13:02, KA2WEU--- via time-nuts wrote:
> You guys never give up, happy Sunday
> 
> 
> In a message dated 10/9/2016 2:46:02 P.M. Eastern Daylight Time,
> magnus at rubidium.se writes:
> 
> Hi,
> 
> Agree. However, one need to make sure that the counter  triggering 
> never
> flukes a measurement.
> 
> There is a few things  missing to make it work much much better.
> 
> Cheers,
> Magnus
> 
> On  10/09/2016 08:35 PM, Bob Camp wrote:
>> Hi
>> 
>> I understand  the “keep it simple” concept, even if I rarely practice 
>> it
> :)
>> 
>>  I would indeed like to get time tagging of phase measurements better
> integrated with some of these
>> tools. The whole “was that a dropout in  the signal or a counter 
>> issue”
> thing is rarely handled in a
>> very good  fashion. It also just happens to be a pretty good addition 
>> to
> a comb  measurement system
>> as well.
>> 
>>  Bob
>> 
>> 
>>> On Oct 9, 2016, at 1:33 PM, Magnus Danielson
> <magnus at rubidium.dyndns.org> wrote:
>>> 
>>> Hi  Bob,
>>> 
>>> There is so many things that could be done  differently if we started
> with a clean sheet. I was intentionally not going  down that road but 
> more
> thinking about practical setups with the stuff we  have, or very small
> additions.
>>> 
>>> Cheers,
>>>  Magnus
>>> 
>>> On 10/09/2016 07:26 PM, Bob Camp  wrote:
>>>> Hi
>>>> 
>>>> 
>>>>>  On Oct 9, 2016, at 1:22 PM, Magnus Danielson
> <magnus at rubidium.dyndns.org>  wrote:
>>>>> 
>>>>> Hi Bob and  Bob,
>>>>> 
>>>>> This is why the two-counter setup  is so messy, you have to have
> software that will sync up and query them  alternatively. You also need 
> to make
> sure you get the counters to trigger.  Besides, another issue is that
> difference in the two counters read-outs will  cause a false signal,
> so calibration
> and compensation becomes important to  remove that.
>>>> 
>>>> That’s why I believe the time  tagger + counter is the better 
>>>> solution
> rather than multiple counters. Let it  give you the global information 
> and
> then use it to sort out what you see from  the counter. Yes, a full 
> blown
> multi channel time tagger with picosecond  resolution would be better 
> still.
> That’s going to cost more than  $5….
>>>> 
>>>>  Bob
>>>> 
>>>>> 
>>>>> Using a picket  fence type of triggering approach is cheaper and
> easier to maintain. Some mild  software support for the processing and 
> it will
> work like a charm. Calibration  for true zero offset is needed, but 
> relatively
> easy to implement, you want  that anyway.
>>>>> 
>>>>>  Cheers,
>>>>> Magnus
>>>>> 
>>>>> On  10/09/2016 07:02 PM, Bob Stewart wrote:
>>>>>> Hi  Bob,
>>>>>> I had actually thought about making a server for  the Prologix
> Ethernet adapters, but I gave up when I considered the issue of  two 
> processes
> trying to claim the same device.  I've experimented with  using a C 
> program to
> capture multiple GPIB ports to a live file.  But, I  can't figure out 
> how to
> get the "live" part to work when running Timelab on a  Windows client 
> in a
> Virtual Box under a Linux server that is collecting the  data.  I think
> Santa may have to bring me another GPIB adapter this  Christmas.
>>>>>> 
>>>>>>  Bob
>>>>>>  -----------------------------------------------------------------
>>>>>>  AE6RV.com
>>>>>> 
>>>>>> GFS GPSDO  list:
>>>>>>  groups.yahoo.com/neo/groups/GFS-GPSDOs/info
>>>>>> 
>>>>>>   From: Bob Camp <kb8tq at n1k.org>
>>>>>> To:  Bob Stewart <bob at evoria.net>; Discussion of precise time and
> frequency  measurement <time-nuts at febo.com>
>>>>>> Sent: Sunday,  October 9, 2016 11:50 AM
>>>>>> Subject: Re: [time-nuts]  TimeLab
>>>>>> 
>>>>>>  Hi
>>>>>> 
>>>>>>> On Oct 9, 2016, at  12:27 PM, Bob Stewart <bob at evoria.net>  
>>>>>>> wrote:
>>>>>>> 
>>>>>>> Hi  Bob,
>>>>>>> Is it actually possible to address two  devices on one GPIB 
>>>>>>> adapter
> with Timelab?  I admit to not reading the  documentation carefully, but 
> I've
> not been able to do this directly.  The  only way I could think of 
> doing it
> was to use some software to send the data  to a file and then use 
> Timelab
> to pull the data from the file.  Maybe NI  software allows you to 
> configure
> this?
>>>>>> 
>>>>>> That was my poorly  stated point :) … you would have to add the
> ability to identify and address  multiple devices.
>>>>>> 
>>>>>>  Bob
>>>>>> 
>>>>>>> 
>>>>>>>  Bob
>>>>>>>  
>>>>>>> -----------------------------------------------------------------
>>>>>>>  AE6RV.com
>>>>>>> 
>>>>>>> GFS GPSDO  list:
>>>>>>>  groups.yahoo.com/neo/groups/GFS-GPSDOs/info
>>>>>>> 
>>>>>>>   From: Bob Camp <kb8tq at n1k.org>
>>>>>>>  To: Discussion of precise time and frequency measurement
> <time-nuts at febo.com>
>>>>>>> Sent: Sunday, October  9, 2016 8:42 AM
>>>>>>> Subject: Re: [time-nuts]  TimeLab
>>>>>>> 
>>>>>>>  Hi
>>>>>>> 
>>>>>>> Given that *some*  of us have more than errr … one counter  :)
>>>>>>> 
>>>>>>> There are several  setups that involve two or three counters to
> resolve some of these issues.  Having
>>>>>>> multiple serial ports or multiple devices  on a GPIB isn’t that 
>>>>>>> big
> a problem. Addressing multiple  devices
>>>>>>> (setting up the addresses in TimeLab) is  an added step. Coming 
>>>>>>> up
> with standard setups would be  the
>>>>>>> first step. Getting them documented to the  degree that they 
>>>>>>> could
> be run without a lot of hassle would  be
>>>>>>> the next  step.
>>>>>>> 
>>>>>>> Another fairly  simple addition (rather than a full blown 
>>>>>>> counter)
> would be some sort of MCU  to time tag
>>>>>>> the input(s). It’s a function that is  well within the 
>>>>>>> capabilities
> of a multitude of cheap demo cards. Rather  than
>>>>>>> defining a specific card, it is probably  better to just define a
> standard message (115200 K baud, 8N1,  starts
>>>>>>> with “$timenuts$,1,”, next is the channel  number, after that the
> (32 bit?) seconds count.The final data field  is
>>>>>>> a time in nanoseconds within the second, *two  byte check sum is
> last, cr/lf). If there is a next generation version that  is
>>>>>>> incompatible, the 1 after timeouts changes to a  2.) Yes, even 10
> seconds after typing that definition I can  see
>>>>>>> a few problems with it. Any structural  similarity to NMEA is 
>>>>>>> purely
> intentional. That’s why it needs a bit  of
>>>>>>> thought and work before you standardize on it.  It still would be 
>>>>>>> a
> cheap solution and maybe easier to  integrate
>>>>>>> into the software than multiple  counters. You do indeed have all
> the same setup and documentation  issues.
>>>>>>> 
>>>>>>> In any of the  above cases, the only intent of the added hardware 
>>>>>>> is
> to get a number that is  good to 10’s of ns.
>>>>>>> Anything past that is great.  Once you know where all the edges
> really are, sorting out the phase data  becomes
>>>>>>> much  easier.
>>>>>>> 
>>>>>>>  Bob
>>>>>>> 
>>>>>>>> On Oct 9,  2016, at 7:32 AM, Magnus Danielson
> <magnus at rubidium.dyndns.org>  wrote:
>>>>>>>> 
>>>>>>>> Fellow  time-nuts,
>>>>>>>> 
>>>>>>>> I  don't know if it is me who is lazy to not figure TimeLab out
> better or if it  is room for improvements. I was considering writing 
> this
> directly to John, but  I gather that it might be of general concern
> for many, so
> I thought it be a  good topic for the  list.
>>>>>>>> 
>>>>>>>> In one  setup I have, I need to measure the offset of the PPS as 
>>>>>>>> I
> upset the system  under test. The counter I'm using is a HP53131A, and 
> I use
> the time-interval  measure. I have a reference GPS (several actually) 
> which
> can output PPS, 10  MHz, IRIG-B004 etc. In itself nothing  strange.
>>>>>>>> 
>>>>>>>> In  the ideal world of things, I would hook the DUT PPS to the
> Start (Ch1) and the  reference PPS to the Stop (Ch2) channels. This 
> would give
> me the propper Time  Error (DUT - Ref) so a positive number tells me 
> the DUT
> is ahead of the  reference and a negative number tells me that the DUT 
> is
> behind the  reference.
>>>>>>>> 
>>>>>>>>  Now, as I do that, depending on their relative timing I might 
>>>>>>>> skip
> samples,  since the counter expects trigger conditions. While TimeLab 
> can
> correct for  the period offset, it can't reproduce missed  samples.
>>>>>>>> I always get suspicious when the time  in the program and the 
>>>>>>>> time
> in real world does not match  up.
>>>>>>>> 
>>>>>>>> I could  intentionally shift the PPS output of my DUT to any
> suitable number, which  would be one way to solve this, if I would
> tell TimeLab to
> withdraw say 100  ms. I might want to do that easily afterhand rather 
> than
> in the setup  window.
>>>>>>>> 
>>>>>>>> To  overcome this, I use the IRIG-B004 output, which is a 100 Hz
> signal with a  stable rising edge aligned to the PPS to within about 2 
> ns.
> Good enough for my  purpose. However, for the trigger to only produce
> meaningful results, I will  need to swap inputs, so that the PPS from 
> DUT is on
> Start/Ch1 and the IRIG-B  is on Stop/Ch2. This way I get my triggers 
> right.
> However, my readings have  opposite sign. I might have forgotten about
> the way to
> correct for  it.
>>>>>>>> 
>>>>>>>> However,  TimeLab seems unable to unwrap the phase properly, so 
>>>>>>>> if
> I have the condition  where I would get a negative value of say -100 ns 
> then
> the counter will  measure 9,999,900 ns, so I have to force a positive 
> value
> as I start the  measurement and then have it trace into the negative. I
> would very much like  to see that TimeLab would phase-unwrap into +/- 
> period/2
> from first sample.  That would be much more  useful.
>>>>>>>> 
>>>>>>>> I  would also like to have the ability to set an offset from 
>>>>>>>> which
> the current  zoom window use as 0, really a form variant of the 0-base 
> but
> letting me  either set the value or it be the first value of the zoom. 
> I have
> use for both  of these. I often find myself fighting the offset issues. 
> In
> a similar  fashion, I have been unable to change the vertical zoom, if 
> I
> don't care about  clipping the signal then it forces me to zoom in 
> further than
> I like to. The  autoscale fights me many times in a fashion I don't  
> like.
>>>>>>>> 
>>>>>>>> OK, so  there is a brain-dump of the last couple of weeks on and
> off measurement  experiences. While a few things might be fixed in the 
> usage,
> I wonder if there  is not room for improvements in the tool. I thought 
> it
> better to describe what  I do and why, so that the context is  given.
>>>>>>>> 
>>>>>>>>  Cheers,
>>>>>>>> Magnus
>>>>>>>>  _______________________________________________
>>>>>>>>  time-nuts mailing list -- time-nuts at febo.com
>>>>>>>>  To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>>>>>  and follow the instructions  there.
>>>>>>> 
>>>>>>>  _______________________________________________
>>>>>>>  time-nuts mailing list -- time-nuts at febo.com
>>>>>>> To  unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>>>>  and follow the instructions  there.
>>>>>>> 
>>>>>>> 
>>>>>>> 
>>>>>>>  _______________________________________________
>>>>>>>  time-nuts mailing list -- time-nuts at febo.com
>>>>>>> To  unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>>>>  and follow the instructions  there.
>>>>>> 
>>>>>> 
>>>>>> 
>>>>>>  _______________________________________________
>>>>>>  time-nuts mailing list -- time-nuts at febo.com
>>>>>> To  unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>>>  and follow the instructions there.
>>>>>> 
>>>>>  _______________________________________________
>>>>> time-nuts  mailing list -- time-nuts at febo.com
>>>>> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>>  and follow the instructions there.
>>>> 
>>>>  _______________________________________________
>>>> time-nuts  mailing list -- time-nuts at febo.com
>>>> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>> and  follow the instructions there.
>>>> 
>>>  _______________________________________________
>>> time-nuts mailing  list -- time-nuts at febo.com
>>> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow  the instructions  there.
>> 
> _______________________________________________
> time-nuts  mailing list -- time-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the  instructions there.
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to 
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.

-- 
Dr. Don Latham