[time-nuts] HP5065A +20V supply and C-field mods
Bob kb8tq
kb8tq at n1k.org
Wed Aug 9 14:48:47 EDT 2017
Hi
Sounds like a lot of fun getting everything to zero out :)
One thing to watch for is hysteresis. The tempco may not be quite the same
going hot to cold as going cold to hot.
Bob
> On Aug 9, 2017, at 2:34 PM, cdelect at juno.com wrote:
>
> HP5065A +20V supply and C-field circuit mods
>
> I'm doing some more work on the 5065A and thought I'd share the results
> of a
> modification (one of several that will be done on this unit).
> I measured the +20V P/S temperature coefficient and got -13.5ppm per
> degree C.
> This equates well with the 1N938 10ppm/C and the voltage set resistor TCs
>
> of plus/minus 10ppm/C. However the TC varies if the C-field pot is
> changed.
> The original circuit has 11.9ma thru the reference diode series resistor
> (925 Ohm)for a diode current of 6.8ma and a C-field pot circuit current
> of
> 5.1ma at midrange on the pot. (zero TC on the diode is specified as
> 7.5ma)
> The diode current ranges from 5.3 to 7.8ma depending on the C-field pots
> position.
> Since I am replacing the C-field circuit, the C-field pot circuit has
> been
> removed. The 925 Ohm resistor was replaced with a 1ppm/C 1K resistor and
> a
> 1K precision pot. This allows the Reference diode current to be adjusted
> to compensate for all the elements in the circuit. (The TC of the diode
> changes depending on the diode current)
> After adjusting the diode current and temperature over many iterations
> and monitoring the +20 Volt output I was able to find a diode current to
> bring the +20 Volt output coefficient to below 1ppm per degree C!
> (-.6ppm/C)
> So that part of the mod is done.
> Next I will be installing a new C-field current source that has a TC of
> 2ppm
> per degree C.
> Then I'll do some long term plots against the Maser to see if the long
> term stability has improved at all.
>
> More to come!
> Cheers,
>
> Corby
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the time-nuts
mailing list