[time-nuts] Performance verification for time counters
Bob kb8tq
kb8tq at n1k.org
Wed Nov 29 17:20:26 EST 2017
Hi
> On Nov 29, 2017, at 5:16 PM, Andy ZL3AG via time-nuts <time-nuts at febo.com> wrote:
>
>
> HP 5359A Time Synthesiser?
100 to 200 ps jitter …. he’s after <5 ps
Bob
>
>
> On 30/11/2017, at 10:24 AM, Leo Bodnar wrote:
>
>> I am looking for an established and widely accepted procedure for verifying performance of high resolution time counters.
>>
>> I have designed a time stamping counter for qualifying 1PPS signal performance against external reference (e.g. 10MHz master clock.)
>>
>> Simple design verification check I am doing at the moment is gating random selection of master clock edges back into device's signal input and letting the device measure this test signal offset against its reference clock - which, for ideal design, should result in zero offset (modulo 100ns.) My results are roughly in line with what I expect to see http://leobodnar.com/balloons/NTP/time-sampler-test1.png
>>
>> Now, what would be recognised procedure for sweeping external input pulse delay over few hundred ns in a controlled, measurable and repeatable way?
>>
>> I can see few naïve approaches:
>> 1) Using selectively gated (or divided) reference clock followed by adjustable delay line. E.g. something like mechanically adjusted delay lines used in HP test sets. Or, perhaps, calibrated rigid coax sections?
>> 2) Slightly offset another master clock (e.g. second Rb oscillator) gated in a similar way but without delay line, followed by statistical data analysis
>> 3) Trusted pulse generator with high resolution delay adjustment fed from the same master clock as the counter
>>
>> I am looking for something with ~10ps accuracy, 100ns+ range, and reasonably low jitter (~5ps or better.)
>> It is possible that the range needs to be split up (e.g. fixed rigid coax delay line followed by a mechanically adjust section.)
>>
>> This is a low budget fun project so something simple and common sense is preferred to "price on application" NIST traceable equipment.
>>
>> Thanks!
>> Leo
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the time-nuts
mailing list