We'll, asking the 3457A to read 7.5 digits - exceeds the capability of
the internal reference. A 6.5 digit meter needs a reference to have 0.5
ppm stability. The service manual rates the instrument at 5 PPM, and
when I had a cal lab to test my unit, it came back as 5.8 ppm.
Something else to remember, is the last digit of the multimeter should
not really be used, if your concerned about accuracy.
A good application note to read on this issue is AN82-1 from Linear
Technology. There is a graph on the front page that sums it up.
Mark Sims wrote:
> My main gripe with the 3457A is that high-res mode. It was just plain stupid on HP's part to not make the meter operate directly in 7.5 digit mode. Why does one have to read a separate register and manually add it to the reading? Come now... can't a microprocessor do that better than a person. In fact I built a 3457A GPIB interface to a MegaDonkey touch screen LCD controller specifically to read the high res register and do the addition/display.
>
> Another thing to watch out for on the 3457A is the "seams" between the ranges. Even on units freshly checked out and calibrated from HP I have seen rather large errors in readings when I put a precision resistance decade box or voltage cal source on the thing. Stepping the value from say 2.9V to 3V and from 3.1V to 3V produces different readings that were out of spec. (the 3V point seems the most error prone).
>
> ----------------------------------------
>
>
> _________________________________________________________________
> Get back to school stuff for them and cashback for you.
> http://www.bing.com/cashback?form=MSHYCB&publ=WLHMTAG&crea=TEXT_MSHYCB_BackToSchool_Cashback_BTSCashback_1x1
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
>
More information about the volt-nuts
mailing list