[volt-nuts] Fluke 845AB Null Detector
Alan Scrimgeour
scrimgap at blueyonder.co.uk
Fri Mar 19 20:04:16 UTC 2010
Oops! Forgot to check the manual.
Your memory is correct.
Mine meets the noise spec. but the zero control covers too large a range and
consequently it's impossible to turn it in small enough increments to zero
perfectly.
This is what the 845AB manual says about noise (and zero):
------------------------------------------------------
MAXIMUM NOISE (input shorted)
RANGE NOISE (peak-to-peak)
1 µV 0.20 µV
3 µV 0.25 µV
10 uV - 1000 V 0.30 µV
STABILITY OF ZERO
Better than 0.15 µV/hour
Better than 0.3 µV/day
TEMPERATURE COEFFICIENT OF ZERO
Less than 0.1 µV/°C from 15° C to 35°C
Less than 0.2 µV/°C from 0° C to 50°C
ZERO CONTROL RANGE
±5 µV minimum
------------------------------------------------------
Alan
----- Original Message -----
From: "Mark Sims" <holrum at hotmail.com>
To: <volt-nuts at febo.com>
Sent: Friday, March 19, 2010 6:35 PM
Subject: [volt-nuts] Fluke 883AB differential voltmeter
>
> The noise spec on the 845 is 0.2 uV (if I remember correctly). Mine seems
> to be mostly within 0.1 uV.
>
> As far as battery replacement, I still used nicads (since I had them).
> One advantage of nicads over nimh cells is they are more robust. They
> tolerate overcharge better and have much better "shelf life".
>
> On a lot of replacements I do use the low self discharge NiMH cells.
> Their self discharge rate does seem to be greater than advertised...
>
> For circuits that trickle charge nicads (at say C/12), replacement with
> NiMH cells will give a C/36 charge rate (assuming three times the nicad
> capacity). This is OK, particularly if you don't leave them on charge
> permanently.
>
> Speaking of overcharging NiMH cells... I put a dead 3200 maH 6 cell pack
> on a constant 1.5 A charge and went to the local surplus store... no
> problem... back in a couple of hours. Got stuck in a nasty traffic jam,
> got back in 3.5 hours. The pack had exploded and let the magic smoke out
> of one of the cells... actually magic carbon dust. Parts of the pack
> were scattered over a 10 foot radius. Nasty black powder everywhere...
> parts of pack stuck to walls... etc...
> _________________________________________________________________
> The New Busy is not the old busy. Search, chat and e-mail from your inbox.
> http://www.windowslive.com/campaign/thenewbusy?ocid=PID27925::T:WLMTAGL:ON:WL:en-US:WM_HMP:032010_3
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
More information about the volt-nuts
mailing list