[volt-nuts] 3458A - calibration constants coding
    Poul-Henning Kamp 
    phk at phk.freebsd.dk
       
    Sat Nov  5 11:34:01 UTC 2011
    
    
  
In message <4EB50F9C.7000602 at freenet.de>, Frank Stellmach writes:
>The content of the first two constants by MEMREAD:
Which are the exact addresses you did MREAD from ?
You have understood that MREAD reads 16 bits so you should only
give it even addresses and that for the CALRAM only 8 of the 16
bits are valid ?
	ID?
	HP3458A
	MREAD 393216
	16569
	MREAD 393218
	-7239
	MREAD 393220
	-30791
Which means that the first three bytes are 0x40, 0xe3 and 0x87
	>>> "%x" % 16569
	'40b9'
	>>> "%x" % (65536-7239)
	'e3b9'
	>>> "%x" % (65536-30791)
	'87b9'
	>>> 
>This is my question: Is the content of the NVRAM, if read out on a 
>programmer, also encrypted as described above, or does this occur only 
>by the MEMREAD function?
There is no encryption or anything of the sort.
-- 
Poul-Henning Kamp       | UNIX since Zilog Zeus 3.20
phk at FreeBSD.ORG         | TCP/IP since RFC 956
FreeBSD committer       | BSD since 4.3-tahoe    
Never attribute to malice what can adequately be explained by incompetence.
    
    