[volt-nuts] Integrator errors from dielectric soakage?
Andreas Jahn
Andreas_-_Jahn at t-online.de
Tue Sep 6 21:28:05 UTC 2011
> Any ideas as to how much error the (presumably) Mylar integrator cap
> will introduce? In other words, will such error be significant in a
> mere 3.5 digits?
Hello,
several years ago I built a integrating ADC.
2.5V Reference as "zero voltage" and integrating +5V and 0V
against the input signal.
(0 .. 5V measured as -2.5 to 2.5 V).
Due to capacitor soaking I had a gap in the middle of the 0 to 5V
range which was about 1mV (0.02%) as non linear jump.
I could reduce the non linearity by replacing the mylar capacitor with
a polypropylene capacitor by a factor of around 10.
So for a meter with 3,5 digits this should not play a role.
With best regards
Andreas
More information about the volt-nuts
mailing list