I made it through the DC Cal steps this evening and I was reading about the
Ohms calibration.
What resistance standards are needed? 10000 ohms and 1000 ohms? 4 wire or
2?
Thanks,
Joe
-----Original Message-----
From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces at febo.com] On
Behalf Of Jean-Louis Noel
Sent: Friday, June 22, 2012 9:25 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Calibration Sources for Solartron 7081
Hi J. L.
From: "J. L. Trantham" <jltran at att.net>
> However, AC promises to be a problem. I guess I will try using a
> signal generator and an amplifier to see what output level and
> stability I can achieve.
You have to link the amplifier to your DC standard to have the stability
needed.
> But still, what frequency? 60 Hz? 1 kHz?
The 'basic' calibration is usually performed at 400Hz.
With check at higher frequencies (40KHz and 400KHz for the 7081) to know if
it's still properly compensated.