[volt-nuts] Thermistor value for 732A Reference 10V Standard
Mitch Van Ochten
mitch at vincentelectronics.com
Fri Jun 29 11:03:20 UTC 2012
I think monitoring the 1V or 1.1V is a better test over temperature than
monitoring the 10V. It's quite stable even without the heater.
----- Original Message -----
From: "Rob Klein" <rob.klein at smalldesign.nl>
To: "Discussion of precise voltage measurement" <volt-nuts at febo.com>
Sent: Friday, June 29, 2012 6:29 AM
Subject: Re: [volt-nuts] Thermistor value for 732A Reference 10V Standard
Op 29-6-2012 6:34, Tom Leedy schreef:
> Has anyone actually measured the internal temperature of the oven (in
> degrees
> C)? Since the operating temperature of the 732A is spec'd at 40 C, I
> suspect
> that it is about 45-50 C, or so.
According to the manual it's 48 ± 2°C
> And I also suspect that temperature stability
> of the oven is more important than absolute accuracy.
You'd be correct :-)
>
> I didn't expect the large outpouring of knowledge!
*Nobody* expects the large outpouring of knowledge! (with apologies to
M. Python)
Rob.
_______________________________________________
volt-nuts mailing list -- volt-nuts at febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
More information about the volt-nuts
mailing list