Replying to my own post - but I do have highly pertinent information:
I received an email from a person I know at Ametek:
>As for the lower frequencies, did you try using the AC filter button?
>I tried my (annually calibrated) 7081 on a Fluke 5720A before leaving
>work and at 50Hz I got virtually the same reading as you, 1.00367.
>Selecting the AC filter gave a much improved reading of 1.00002.
>I haven't got the operating manual to hand but I seem to remember it
>recommending using the filter at 50Hz and below.
He was pretty much on the mark, though the manual recommends the filter
be used below 1kHz!
With AC filter engaged:
Frequency Reading Error
50Hz 0.999996 .0004%
100Hz 0.999868 .0132% <--- A little high but not bad
The 500kHz and 1MHz figures are also within specification for the 7081
Cheers
Dave
-----Original Message-----
From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces at febo.com] On Behalf Of David C. Partridge
Sent: 20 March 2012 15:48
To: 'Discussion of precise voltage measurement'
Subject: [volt-nuts] Fluke 5200A AC Calibrator
Checking the output is against my 7081, here is what I'm see on the 1V range set for 1V (the other ranges show a very similar pattern):
Frequency Reading Error
50Hz 1.00371 .371% <--- Surprisingly high
100Hz 1.00091 .091% <--- Surprisingly high