[volt-nuts] hp 3458a advice
Paul Fox
pgf at foxharp.boston.ma.us
Tue Oct 16 15:31:49 UTC 2012
thanks joe, and to the other couple of people who responded privately.
i need to figure out how much of a project i want this to be, and how
it ranks against all the other projects that i also don't have time
for. :-)
i have the full users manual (two copies!), but i don't have anything
else that speaks GPIB. i'll play with some of the front panel
commands you've suggested, while i think about what to do. thanks!
paul
j. l. trantham wrote:
> Paul,
>
> First, the meter is worth something if nothing other than parts. It is
> quite a meter and you should at least open it up and see what it might take
> to get it operational again. I am sure there are folks on this list
> (including me) that would love to have the meter if you want to part with
> it.
>
> It is still in production, still supported and you can get the manuals from
> the Agilent website. The assembly level repair manual should be able to
> give you an idea of where the problem is. The CLIP is available and you
> might be able to find the specific part (or parts) that need replacing.
>
> I don't know of a way to get the serial number if there is no sticker on the
> back. You can open it and look at the date codes on the chips, etc., and
> get an idea of when it was manufactured.
>
> You can enter some queries from the front panel to get some other
> information. I can't remember what they all are but if they end in a '?',
> it will return the data. Things like 'OPT?' will tell what options are
> installed, '1' being extended memory and '2' being the high stability
> reference. You can tell how many times it was CAL'd, what version of
> firmware it has, etc. I think version 9 is the latest and version 8 is 'out
> there' somewhere, IIRC. If you just start with shift A and scroll down the
> list, you can see what questions you can ask.
>
> Agilent offers a 'repair service' for about $1900 (IIRC) that will put it
> back in operational condition, up to specs, and with a fresh calibration.
> However, I don't know if that service requires a serial number or not.
>
> There is a guy at the Agilent calibration center in Colorado, Gary Bierman,
> and he is 'the man' when it comes to that meter. I have spoken with him in
> the past about other 3458A questions and he was most helpful. I suspect he
> would be more than willing to talk to you about the meter and what it might
> take to get it up and running.
>
> I think his contact information is in the archives if you would like to
> pursue that.
>
> Good luck.
>
> Joe
>
> -----Original Message-----
> From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces at febo.com] On
> Behalf Of Paul Fox
> Sent: Monday, October 15, 2012 9:50 AM
> To: Discussion of precise voltage measurement
> Subject: [volt-nuts] hp 3458a advice
>
> hi --
>
> i'm not a volt-nuts regular, and not even a precision measurement nut
> at all, but given past discussion in the list archives, i think you
> folks might be able to help.
>
> i was recently given an HP 3458a multimeter. the donor thought it
> worked when he last used it, but that was quite a long time ago, and i
> think it was stored in an attic for some time, so i'm not too
> surprised that it now doesn't pass self tests. :-/ he mainly wanted
> it out of his house, along with a bunch of other vintage (sadly mostly
> non-working) equipment he'd accumulated over the years.
>
> the self-test error is:
> ERRSTR 204, "Hardware Failure -- Flatness DAC Convergence: 198"
>
> judging from past list traffic, repairs on these units are
> expensive, and it would still need recalibration to be useful, and
> there are may be other parts (eeproms?) that may not be good
> any more. i'm not really interested in pursuing any of this
> myself.
>
> what's worse is that the unit doesn't have a serial number. (i assume
> the s/n would normally be obvious on the rear of the unit.) it bears a
> sticker on the back that says "MTG PROTO 27". i think the unit came
> from a place that themselves produced reference standards, and they
> may have gotten a very early model from HP to work with. that's
> conjecture. i haven't opened the unit, so i don't know if there are
> other clues to its age inside.
>
> so: is this machine useful to anyone? given the error message above,
> are there any measurements at all that it might still get correct, in
> it's current state?
>
> paul
> =---------------------
> paul fox, pgf at foxharp.boston.ma.us (arlington, ma, where it's 63.1 degrees)
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
=---------------------
paul fox, pgf at foxharp.boston.ma.us (arlington, ma, where it's 48.4 degrees)
More information about the volt-nuts
mailing list