[volt-nuts] Calibration Device
J. Forster
jfor at quikus.com
Mon Aug 12 12:09:54 EDT 2013
My main interest is as a 'sanity check'. I use DMMs for engineering purposes.
Shields up:
IMO, there are very few applications, other than fundamental physics
research, that really NEED 5,6,7, or 8+ digits.
I would no more use a multi digit DVM to do what can be done with a
differential voltmeter than use a carrier phase tracking GPS to measure
for a living room carpet.
Shields down.
-John
================
> John wrote:
>
>>David mentioned this:
>> * * *
>>What do people think of the device?
>
> What are you trying to accomplish? As I posted previously:
>
>>Let's assume that it is still working exactly the same as it was
>>when it was calibrated. The calibration values are recorded to
>>4-1/2 digits. So the uncertainty is greater than the two LSDs of a
>>3457A. Of course, it is almost certainly not working exactly the
>>same as it was when it was calibrated, [so its last digit is going
>>to be uncertain,] which is going to put at least the third LSD of a
>>3457A in question. So it is good enough for a 3457A only if one is
>>content with a 3-1/2 digit verification on a 6-1/2 digit instrument.
>
> You should be able to rely on it to verify instruments to 3-1/2
> digits. Beyond that, you would need something with a smaller uncertainty.
>
> Best regards,
>
> Charles
>
>
>