[volt-nuts] HP 3457A

Mike S mikes at flatsurface.com
Mon Aug 12 15:07:03 EDT 2013


On 8/12/2013 2:23 PM, John Phillips wrote:
> A calibration indicates that the unit under test is withing manufactures
> specification. The equipment and procedure used has to be "good enough"
> (bad words in a cal lab) to have a high probability (nothing is 100%) of
> insuring the calibration documentation is valid. Things can  can be a
> little looser if you are calibrating a 1% meter with a 10 ppm meter but it
> does not work the other way around.

You can calibrate either way. You can't however, calibrate the 10 ppm 
meter so it's in spec using a 1% meter. That's different. Calibration 
merely means that it's documented how close it is to a reference, such 
as NIST, not that it's within the manufacturer's spec. The 10 ppm meter 
would end up with a 1%+ calibration - precise but not accurate. Not 
particularly useful, but valid. A good cal lab would do a calibration to 
specification, where the uncertainties place the 10 ppm meter within spec.

As I cited and someone else already quoted, calibration is the "property 
of a measurement result whereby the result can be related to a reference 
through a documented unbroken chain of calibrations, each
contributing to the measurement uncertainty." Nothing to do with making 
a device meet its specifications.

That's why an eBay seller can claim they'll do a "calibration traceable 
to NIST," because they're not claiming any particular accuracy. It's 
really not worth anything, unless they give specific uncertainties or 
claim calibration to manufacturer's specification.