Hi ,
Newer does not mean better. It's not "nuts" level, but my Fluke 8060A 4.5 digit handhelds are more accurate and stable than my newer 89 IV. However most "engineers" would pick up the 89.
I agree with Dave on UK " engineers" For many years I've been a member of the very few groups of engineer in the UK who need a licence to do their job - aircraft, but even that has been diluted since EU regulations came into force. I'm also a Chartered Engineer similar to PE in the USA. As I understand it most US States require you to be a PE before you can do business as an "Engineer". In the UK it's a bloke with an oily rag and big hammer :-(
 
Robert CEng MRAeS G8RPI 


________________________________
From: Orin Eman <orin.eman at gmail.com>
To: Discussion of precise voltage measurement <volt-nuts at febo.com> 
Sent: Tuesday, 13 August 2013, 5:56
Subject: Re: [volt-nuts] Calibration and Certification - Trust and detail


On Sun, Aug 11, 2013 at 3:36 PM, Orin Eman <orin.eman at gmail.com> wrote:

>
> I should go get the kelvin clips out and compare the 3456A against the
> '61A on some 10k precision wirewound resistors I have.
>


I did.  The resistors are MR102 series 0.01% 1/8W wirewound:

34461A: 10.000 82 +/- 0.000 90
3456A: 10.000 98 +/- 0.000 58

Using 90 day specs for the 61A and 90 day plus 0.0004% per month for the
56A.  Yes, the 3456A resistance specs are better than the shiny new 34461A.

I also compared the DC voltage ranges from 0.1 to 1000V and other than on
the 1KV range, results were within 10ppm.  Still, given the 3456A spec
sheet says add .12(input voltage/1000)^2 % on the 1KV range, I can't
complain; I got 999.984 on the 61A and 1000.062 on the 56A.  I'm pretty
sure that the superscript 2 in the spec meant squared, not that it matters
for 1000/1000.

Now has anyone calibrated a Fluke 343A?  The instructions in the manual are
entertaining.  They tell you to adjust the 1000V range to +/- 1mV, but the
post calibration check says that a freshly calibrated unit should be +/-
100uV at the 1000V setting.

Orin.