Daniel,
They are made like that... Problem is with drift.
When you cal a 3458A the first step is to short the inputs and wait for the
thermals to die. Then you do a Cal 0 and the incitement stores all the 0
offsets for that set of terminals. they you switch to the other set and do
it again.
Next step is apply 10 votes to the terminals or in my case it is 9.9999411
and enter cal 9.9999411 front and back set of terminals. remove the voltage
and plug in a 10k standard resistor and in my case enter cal 9999.884 for
front and back. In most cases AC does not have to be done. The meter is
comparing its measure values with the values entered and calculates the
correction factor to be used each time a value is displayed.
Most good meters now days do have a null function so you can look at drift
or compare 2 values.



On Wed, Aug 14, 2013 at 6:17 PM, Daniel Mendes <dmendesf at gmail.com> wrote:

>
> Sorry if i´m being naive, but what´s the difficuty of making a digital
> equipment with a memory to store the offset of each scale ans subtract it
> before sending to the display, no pot trimming involved?
> Why aren´t all of the ones made after, let´s say, 1995, like this?
>
> Daniel
>
> Em 14/08/2013 20:29, Dr. David Kirkby escreveu:
>
>  On 14 August 2013 06:41, Charles Steinmetz <csteinmetz at yandex.com> wrote:
>>
>>> Joe wrote:
>>>
>>>  The way I read this is that if I send them a DMM that is within spec,
>>>> they
>>>> won't adjust it or provide pre/post data. Is this the case? If I spend
>>>> over
>>>> $200 sending a DMM to them, I want it adjusted to the best possible
>>>> specs
>>>> and I want the data. I do not want someone just saying that it is good
>>>> enough and send it back to me. I can get that for $50 in El Paso.
>>>>
>>> The big difference is not between adjusting and not adjusting -- it is
>>> between getting a calibration "with full data" and getting one without
>>> data.
>>> /The true value of calibration is not the adjustment -- it is the data./
>>>
>>> Agilent doesn't just say it is good enough -- they tell you specifically
>>> how
>>> far off it is and quantify the statistical uncertainty of their
>>> measurement.
>>> That is everything you need (i) to correct readings you make with the
>>> instrument and (ii) to be confident of the potential uncertainty of those
>>> measurements.
>>>
>>> Let's say your meter has an uncertainty spec of +/- 15 uV (1.5 ppm)
>>> total at
>>> 10 V.  If your calibration certificate says the meter reads dead on at
>>> 10.000000 V, the reading shown on the display is your measurement result
>>> (with a certainty of 1.5 ppm, or +/- 15 counts from the reading) when you
>>> measure a 10 V source.  But the cal certificate could just as well say
>>> that
>>> the meter reads 10.000008 V when measuring a 10.000000 V source.  In that
>>> case, you know to subtract 0.000008 V from whatever the meter reads when
>>> you
>>> measure a 10 V source to get your measurement result (again, with a
>>> certainty of 1.5 ppm, or +/- 15 counts from the corrected reading).  Of
>>> course, in the real world a voltage standard will have its own
>>> calibration
>>>
>> One of the advantages of modern instruments over older ones is that
>> measurements are often more convenient to make. This can reduce your
>> measurement time and so cost. For many companies, a case can be made
>> to upgrade if a newer instrument will save time and money.
>>
>> As a rough guess, I would assume 99.999% of instruments sold sold by
>> Agilent are for commerical non-metrology work. Those 99.999% of users
>> do not want to remember to subtract 0.000008 V -  they want that
>> instrument to be as accurate as possible.
>>
>> Now if you take an instrument like the Agilent 3458A 8.5 digit DVM,
>> then the intended user base is going to have a lot of metrologists.
>> Those people might prefer their instruments are not adjusted, but I
>> think for 99.999% of users of test equipment, they would want the
>> instruments adjusted. With so much done in software now, arguments
>> about pots drifting once adjusted dont make any sence.
>>
>> By its very nature, the readers of volt-nuts will often fall into the
>> 0.001% that might not want their instruments adjusted, but I think it
>> is fair to say most would.
>>
>> Agilent must have thought about these arguments, and have come to a
>> decision not to adjust. I'm a bit surprised myself, but they obviously
>> have their reasons. Clearly if an adjustment requires someone to go in
>> with a screwdriver, then it takes time, has some element of risk of
>> causing accidental damage, and it might well cause things to drift
>> more in the short term.
>>
>> Dave
>> ______________________________**_________________
>> volt-nuts mailing list -- volt-nuts at febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/**
>> mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>> and follow the instructions there.
>>
>
> ______________________________**_________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/**
> mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
> and follow the instructions there.
>



-- 
John Phillips


More information about the volt-nuts mailing list