[volt-nuts] HP 3458A repair.

John Phillips john.phillips0 at gmail.com
Thu Sep 19 19:47:04 EDT 2013


Thanks for all your comments.

Gary said he had 39 years invested in HP/Agilent. He would like us to all
send our problems to him and he would get them fixed. I would like to do
that as well but $2600 +$1600 for the cal I need gets makes the 99%
calibrated meter book value very low. I do know they need to make money but
I would rather be able to buy the AC board for $1000 than pay the $2600. I
am quite cheep that way.

I have an other 3458A and the data that tells me how this meter was. I have
several 3325As. We should be able to do some before and after readings. I
would like to fix this one but we may end up sending the other 3458A in and
just using this one here. It does not have a sticker but it does show our
cal was withing spec which is an indirect cal check of our standards.

I am sure that all the US line people would rather have the meter line
here. I talked with someone like Gary if not Gary about 2 years ago. I was
having a problem with the reference drift.  He said he was tired of
training people and having them leave. The new crew did not know how to run
the josephson junction.  They also had a problems with the aged reference.
When they met spec they were "harvested" and expected to stay in spec...
(thinking like an MBA) they did not. They have to be burned in and screened
for stability.
They loose stability in the parts bin. They have to be run for a few weeks
to get back to where they were.


On Wed, Sep 18, 2013 at 9:53 PM, Tom Miller <tmiller11147 at verizon.net>wrote:

> OOPs, disregard. I tangled it with another message.
>
>
> Tom
>
> ----- Original Message ----- From: "Tom Miller" <tmiller11147 at verizon.net>
>
> To: "Discussion of precise voltage measurement" <volt-nuts at febo.com>
> Sent: Wednesday, September 18, 2013 11:28 PM
>
> Subject: Re: [volt-nuts] HP 3458A repair.
>
>
>  Just the one from the Agilent site.
>>
>> If I get one of these, I'll hit Dave up for a good copy.
>>
>> Thanks,
>> Tom
>>
>> ----- Original Message ----- From: "J. L. Trantham" <jltran at att.net>
>> To: "'Discussion of precise voltage measurement'" <volt-nuts at febo.com>
>> Sent: Wednesday, September 18, 2013 10:54 PM
>> Subject: Re: [volt-nuts] HP 3458A repair.
>>
>>
>>  Forgot to ask.
>>>
>>> Do you have the CLIP?
>>>
>>> Joe
>>>
>>>
>>> -----Original Message-----
>>> From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces@**febo.com<volt-nuts-bounces at febo.com>]
>>> On
>>> Behalf Of John Phillips
>>> Sent: Wednesday, September 18, 2013 3:51 PM
>>> To: Discussion of precise voltage measurement
>>> Subject: Re: [volt-nuts] HP 3458A repair.
>>>
>>> I talked with Gary... it sounded more like a sales pitch like they
>>> replace
>>> relays before they failed. The date code on the cal ram is 6 years past
>>> replacement.  The only problem seems to be the amplifier/attenuator
>>> flatness. He said this could be caused by caps aging and changing value.
>>> They tried calibrating the AC but it did not come withing spec so they
>>> did
>>> not update the as left data... I think if they are going to send the as
>>> left data it should be real as left or blank. This meter looks like a
>>> good
>>> meter for what I will use it for so we will not use the AC above 2MHz as
>>> if
>>> we ever did.
>>>
>>>
>>>
>>> On Wed, Sep 18, 2013 at 9:17 AM, J. L. Trantham <jltran at att.net> wrote:
>>>
>>>  I agree with your assessment of an 'old' meter being more desirable.
>>>>
>>>> I would recommend a conversation with Gary Bierman if you have not
>>>> already
>>>> done that.
>>>>
>>>> Joe
>>>>
>>>> -----Original Message-----
>>>> From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces@**febo.com<volt-nuts-bounces at febo.com>]
>>>> On
>>>> Behalf Of John Phillips
>>>> Sent: Wednesday, September 18, 2013 10:45 AM
>>>> To: Discussion of precise voltage measurement
>>>> Subject: Re: [volt-nuts] HP 3458A repair.
>>>>
>>>> The meter will be here in a few days. I did buy the ram. As far as the
>>>> meter
>>>> age goes, unless it is a very old hardware revision, I would rather have
>>>>
>>> an
>>>
>>>> old meter because it is more stable. That is until it brakes.
>>>>
>>>> I do have before and after readings and the 4 readings are the only ones
>>>> that failed. The after readings are better than the before readings.
>>>> Because the 8 and 10 MHz failed we did not get a certificate but we do
>>>>
>>> have
>>>
>>>> a good limited calibration up to 2 MHz. Just no official cal.
>>>>
>>>> The cal we wanted was $1600. We will still have to pay that after the
>>>> repair.
>>>> From what I can tell they did not run SCAL or the before/after readings
>>>> would not have matched. Like they really did not run after readings or
>>>> there
>>>> would have been some mismatch.
>>>>
>>>>
>>>>
>>>>
>>>>
>>>> On Tue, Sep 17, 2013 at 6:50 PM, J. L. Trantham <jltran at att.net> wrote:
>>>>
>>>> > John,
>>>> >
>>>> > Can you give us more information?  Serial number, Rev. number, CALNUM?
>>>> > How much to invest will be determined by age and other condition.
>>>> >
>>>> > It doesn't sound like a simple CALRAM issue but changing the CALRAM is
>>>> > relatively easy.  I removed all three DALLAS chips in mine and
>>>> > installed sockets.  The CALRAM can be read with a chip programmer and
>>>> > the data written to a new DALLAS chip.
>>>> >
>>>> > I would also call Gary Bierman at the Loveland Cal Lab and have a long
>>>> > talk with him.  He has a lot of insight into these meters and
>>>> > generally prefers to do a component level repair rather than an
>>>> > assembly level repair.  The charge sounds like their standard repair
>>>> > charge, no matter what the problem is, and includes a 'fresh
>>>> > calibration' along with a warranty, a year I think, but Gary will be
>>>>
>>> able
>>>
>>>> to answer that question.
>>>> >
>>>> > Also, once you get the meter calibrated by Agilent (and thus prove it
>>>> > is functioning normally) it will be eligible for their 'repair
>>>>
>>> agreement'
>>>
>>>> > which
>>>> > is $178.68 per year.  I would consider buying a 5 year agreement after
>>>> > the repair.
>>>> >
>>>> > Good luck.
>>>> >
>>>> > Joe
>>>> >
>>>> > -----Original Message-----
>>>> > From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces@**febo.com<volt-nuts-bounces at febo.com>
>>>> ]
>>>> > On Behalf Of John Phillips
>>>> > Sent: Tuesday, September 17, 2013 3:36 PM
>>>> > To: Discussion of precise voltage measurement
>>>> > Subject: [volt-nuts] HP 3458A repair.
>>>> >
>>>> > Hi,
>>>> > I have a 3458A that we sent to Agilent for calibration which it >
>>>> failed.
>>>> > Before we sent it we calibrated it and it looked good to us. The
>>>> > infor. we revived led us to believe that the cal memory may have >
>>>> caused
>>>> the failure.
>>>> >  We ask that it be sent buck to us and paid half the cal charges
>>>> > (about
>>>> > $800) insted of the $2660.64 they wanted to repair it. We were just
>>>> > going to repalce the ram in try again.
>>>> > When we got the meter back it came with befor and afer data Like
>>>> > before 10 volts read 9.9999957 and after it read 10.00009 so they did
>>>> > something or the meter drifted that much.
>>>> > The problem is  0.1 volt and 1.0 volts failed at 8 and 10 MHz but
>>>> > passed at
>>>> > 4 MHz.
>>>> > 4MHZ 0.1 volt reads 0.097251 Lower Limit is 0.095930 PASSED 8MHZ 0.1
>>>> > volt reads 0.085712 Lower Limit is 0.0959
>>>> > 2
>>>> > 0
>>>> > FAILED
>>>> > 10MHZ 0.1 volt reads 0.75569 Lower Limit is 0.084900 FAILED
>>>> >
>>>> > 4MHZ 1 volt reads 0.97272 Lower Limit is 0.95930 PASSED 8MHZ 1 volt
>>>> > reads
>>>> > 0.86389 Lower Limit is 0.95920 FAILED 10MHZ 1 volt reads 0.73514 Lower
>>>> > Limit is 0.84900 FAILED
>>>> >
>>>> > The AC after readings are the same. I do not see how AC after could be
>>>> > that identical even if they did not try to calibrate it. Did they just
>>>> > copy the before data and call it after data?
>>>> >
>>>> > My best guess is that if the 4 MHz is in and the higher frequencies
>>>> > are not the meter requires some kind of mechanical adjustment to get
>>>> > the
>>>> frequency
>>>> > response   withing spec or the AC board needs to be repaid.
>>>> >
>>>> > Are they charging a standard repair charge to do a calibration? I do
>>>> > not see changing the memory to fix this.
>>>> >
>>>> > Where would you go from here if this was your meter?
>>>> >
>>>> >
>>>> > --
>>>> > John Phillips
>>>> > ______________________________**_________________
>>>> > volt-nuts mailing list -- volt-nuts at febo.com To unsubscribe, go to
>>>> > https://www.febo.com/cgi-bin/**mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>>>> > and follow the instructions there.
>>>> >
>>>> > ______________________________**_________________
>>>> > volt-nuts mailing list -- volt-nuts at febo.com To unsubscribe, go to
>>>> > https://www.febo.com/cgi-bin/**mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>>>> > and follow the instructions there.
>>>> >
>>>>
>>>>
>>>>
>>>> --
>>>> John Phillips
>>>> ______________________________**_________________
>>>> volt-nuts mailing list -- volt-nuts at febo.com
>>>> To unsubscribe, go to
>>>> https://www.febo.com/cgi-bin/**mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>>>> and follow the instructions there.
>>>>
>>>> ______________________________**_________________
>>>> volt-nuts mailing list -- volt-nuts at febo.com
>>>> To unsubscribe, go to
>>>> https://www.febo.com/cgi-bin/**mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>>>> and follow the instructions there.
>>>>
>>>>
>>>
>>>
>>> --
>>> John Phillips
>>> ______________________________**_________________
>>> volt-nuts mailing list -- volt-nuts at febo.com
>>> To unsubscribe, go to
>>> https://www.febo.com/cgi-bin/**mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>>> and follow the instructions there.
>>>
>>> ______________________________**_________________
>>> volt-nuts mailing list -- volt-nuts at febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/**
>>> mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>>> and follow the instructions there.
>>>
>>>
>> ______________________________**_________________
>> volt-nuts mailing list -- volt-nuts at febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/**
>> mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
>> and follow the instructions there.
>>
>
> ______________________________**_________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/**
> mailman/listinfo/volt-nuts<https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts>
> and follow the instructions there.
>



-- 
John Phillips