[volt-nuts] 732A drift
J. L. Trantham
jltran at att.net
Sat Aug 23 07:41:10 EDT 2014
I think the 'IN CAL' LED has to do with a 'minimum voltage' but I don't
think it has to with a minimum battery voltage.
When you remove the battery pack, with the unit plugged in, I don't think
the 'IN CAL' LED goes out. Otherwise, you would never be able to swap
battery packs.
Joe
-----Original Message-----
From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces at febo.com] On
Behalf Of Todd Micallef
Sent: Saturday, August 23, 2014 6:21 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A drift
The output impedance of both the 1v outputs is ~1k ohm. I wonder if Randy
could check the 3458a input impedance with a 10M ohm resistor and the 10v
output of the 732a.
I was thinking that the 732a needed a minimum battery voltage to allow the
in cal led to turn on with the short to the Lo terminal. I guess I could try
it on one of mine that is waiting for repair.
Todd
Sent from my iPad
> On Aug 23, 2014, at 6:42, "J. L. Trantham" <jltran at att.net> wrote:
>
> Randy,
>
> The 'IN CAL' LED is turned on as Todd describes. It goes off if power
> to the unit is lost and, thus, calibration is lost. It is not related
> to the battery charge LED.
>
> As I understand it, the units are designed to powered on 24/7/365 and
> are 'IN CAL' once they are powered up, stable (weeks, months?), and
> have been calibrated by your reference lab. Once power is lost,
> meaning lost AC and batteries depleted, the 'IN CAL' light goes out
> and outputs are thereafter unreliable.
>
> Once you decide to get the 732A calibrated, you will need to find a
> way to ship it to the reference lab and get it shipped back while
> continuously powered, connecting an external battery pack to the
> connector on the back of the battery pack. The internal battery pack
> is likely to last only a few hours.
>
> There are at least two types of connectors for an external battery,
> two 5 way binding posts and a Hypertronics connector which is a small
> black connector about 'dime' sized. I can find the part number for
> the mating connector if you need it.
>
> I think the +/- 1 uV drift with the 10 V output and the 3458A are
> within specs for both the 732A and 3458A.
>
> Not sure what to think about the drift on the other readings unless
> there is some sort of 'dirt' on the 1 V and 1.018 V binding posts and/or
ground.
>
> Good luck.
>
> Joe
>
>
>
> -----Original Message-----
> From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces at febo.com]
> On Behalf Of Todd Micallef
> Sent: Saturday, August 23, 2014 4:13 AM
> To: Discussion of precise voltage measurement
> Subject: Re: [volt-nuts] 732A drift
>
> Randy,
>
> I looked at the 3458a input impedance, and it is difficult to get the
> meter set to the FixedZ (10M) mode. However, the 100v range is 10M. It
> should default to HiZ on reset.
>
> I forgot to mention to check the outputs and guard resistance to
> ground. One of my 732a had some foam baffling under the cover that had
> dry rotted. There was a lot of leakage to ground.
> I can't remember if the foam was on the older or newer versions of the
732a.
>
> The cal light comes on with a short wire stuck in the hole and
> connected to a LO output terminal. I don't know, but maybe the battery
> charge led has to be off.
>
> Todd
>
> Sent from my iPad
>
>> On Aug 23, 2014, at 1:16, Randy Evans <randyevans2688 at gmail.com> wrote:
>>
>> I hope someone can help with a strange anomaly on either my 3458A or
>> the 732A. The 732A 10V output as measured on my 3458A seems
>> relatively stable over time (it bounces around about +/-1 uV but is
>> it the
> 732 or the 3458?).
>> However, the 1.000 VDC output drifts downward at a rate around 1-2 uV
>> per second as soon as I plug the 3458A into the 732 output. If I
>> remove the 3458A and connect it back up after a few 10s of seconds,
>> the reading goes back to what it started at and then drifts downward
>> again. The 1.018V output also drifts downward but at a much slower
>> rate and not as much. Now the question is: is it the 732 or the
>> 3458A? I tried to see it on my Agilent 34401A DMM but it really
>> doesn't have the resolution, but I do seem to see it on the 1.000VDC
>> output. If it is the 732A, what would cause it to drift downward
>> like that? Since the 3458A has an input impedance of
>>> 10Gohm on the 1 and 10 V ranges, I wouldn't think the 732 would even
>>> see
>> the difference of whether the 3458A is connected of not, but it
>> clearly makes a difference as to how long its connected and how long
>> it's been disconnected.
>>
>> Any one have any conjectures?
>>
>> Also, what will turn on the "In Cal" light? What does it mean if it
>> doesn't come on?
>>
>> Thanks,
>>
>> Randy
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts at febo.com To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
_______________________________________________
volt-nuts mailing list -- volt-nuts at febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
More information about the volt-nuts
mailing list