Hello Ken,

Glad to hear the SpecAn is still going well.

I know working on these machines without extender boards is a pain. I 
assume U1 output is moving as you adjust R6. U1 feeds a differential 
cascode circuit consisting of Q3,4,5,and 6. these would have been 
matched originally for minimum offset. It could be they are no longer 
matched close enough for the offset adjustment to compensate. It could 
be U1 also.

If CR3 anode is -31.8V that means Q9 must be cutoff. Probably the 
quickest is to check every transistor in the input and mid stage 
amplifiers. Also check every carbon resistor, at this age they will all 
be out of tolerance, sometimes quite a bit.

Regards


On 09/07/14 09:10, Kgoodhew wrote:
> Hi Gerd,
> 	Thanks for the reply.
> My mistake I should have said r6 and not r26 as I typed it from memory.
> The  voltages that are high are :-
>
> Q2 collector 7.35v
> Q3-Q4 emitters 2.0v-3.2v should be nearly equal and about 0.7 to 1.8v
> Q5 Q7 collectors -38.0v should be +28.4v
> CR3 anode -31.8v should be -20 to -22v
>
> Originally found Q36 and Q37 faulty apparently caused by intermittent bad
> contacts on the plug in base  on Q15 as I could hear arcing from here after
> powering up the unit and found the plastic plug in base melted, the
> transistor tested ok so I just removed the base and soldered it into the
> pcb.
> During my investigations I found and relaced several out of tolerance
> resistors as well as CR19 -CR20 zener's one of which was breaking down well
> below it's 20v rating. Also found Q33 faulty.
> Originally I had to replace a number of components in the power supply
> regulator board including caps and the U2 regulator for the +19v and +5v
> supplies as well as a few tracks that had been damaged by someone's earlier
> repair attempts, to get the unit to operate at all.
> The unit output is pretty close according to my HP3478A, it is just that I
> have this 1mv  offset on the output that I cannot adjust out using R6.
> As I do not have an extender board to do testing with the board out of the
> chassis it has been a long and painful process to remove the board and
> solder wires on to each point to make measurements that I need each time.
> I even tried to get the sockets to make an extender board but it appears the
> 43/86 pin sockets at the required pin spacing are no longer available.
>
> BTW the HP8569B you repaired for me some time ago is still going strong.
>
>
>
> Today's Topics:
>
>     1. Re: Fluke 5200a (Gerd)
>     2. Re: Buying HP-3458A (Todd Micallef)
>     3. Re: Fluke 5200a (John Phillips)
>
>
> ----------------------------------------------------------------------
>
> Message: 1
> Date: Tue, 08 Jul 2014 10:30:31 +1000
> From: Gerd <admin at controlelectronics.com.au>
> To: volt-nuts at febo.com
> Subject: Re: [volt-nuts] Fluke 5200a
> Message-ID: <53BB3BA7.2070102 at controlelectronics.com.au>
> Content-Type: text/plain; charset=ISO-8859-1; format=flowed
>
> Hello Ken,
>
>   From what I remember of this circuit, R26 is for output stage bias. R6 is
> for adjusting offset. By the way, which voltages are high and how much?
>
> The 5200 is an old beast, expect a lot of marginal electrolytics, carbon
> resistors that have drifted well out of tolerance and even transistors that
> no longer perform any where near what they originally did.
>
> Regards
>
> On 07/07/14 14:56, Ken Goodhew1 wrote:
>> Hi all,
>>
>>                   I am currently repairing a Fluke 5200a ac voltage
>> calibrator.
>>
>> After pulling cards in and out and soldering multiple wires on to test
>> I now have the unit working about 98%.
>>
>> Anybody might have or know of where I might be able to obtain an
>> extender card for this unit as it would make it a lot easier to fix
>> the last remaining problem that I have, ie:- I have a dc offset
>> voltage on the output of about 1mv when it should be +/- 100 microvolts.
>>
>> I am unable to adjust this with r26 adjustment on the power amplifier
>> board as described in the manual. I have followed the troubleshooting
>> guides as per the manual, as well as testing just about all the
>> components around that particular part of the circuit.
>>
>> The problem appears to be that the input stage is not being turned on
>> hard enough as all the suggested voltages for the input stage
>> (Q1-2-3-4-5-9) are all too high. The supply voltages are all within spec.
>>
>> Anybody can offer suggestions?
>>
>> Sorry to put this out there in this group but I cannot find an online
>> group related to fluke test equipment to ask.
>>
>> Ken Goodhew.
>>
>>
>>
>> ---
>> This email is free from viruses and malware because avast! Antivirus
> protection is active.
>> http://www.avast.com
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts at febo.com To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>>
>>
> --
> sig.html *Control Electronics Pty Ltd*
> The home of Shiftmaster
> Quality used test equipment
> *Phone*: 02 9456 0736
> *Website*: www.controlelectronics.com.au
> <http://www.controlelectronics.com.au/>*
> Email*: admin at controlelectronics.com.au
> <mailto:admin at controlelectronics.com.au>
> ABN 86 109 858 022
> ------------------------------------------------------------------------
> The email addresses in this message are not to be added to any database,
> list, or mailing list without my prior explicit written permission. This
> email and any files transmitted with it are confidential and intended solely
> for the use of the individual or entity to whom they are addressed. If you
> have received this email in error please notify the system manager. Please
> note that any views or opinions presented in this email are solely those of
> the author and do not necessarily represent those of the company. Finally,
> the recipient should check this email and any attachments for the presence
> of viruses. The company accepts no liability for any damage caused by any
> virus transmitted by this email.
>
>
> ------------------------------
>
>
>
> ---
> This email is free from viruses and malware because avast! Antivirus protection is active.
> http://www.avast.com
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>

-- 
sig.html *Control Electronics Pty Ltd*
The home of Shiftmaster
Quality used test equipment
*Phone*: 02 9456 0736
*Website*: www.controlelectronics.com.au 
<http://www.controlelectronics.com.au/>*
Email*: admin at controlelectronics.com.au 
<mailto:admin at controlelectronics.com.au>
ABN 86 109 858 022
------------------------------------------------------------------------
The email addresses in this message are not to be added to any database, 
list, or mailing list without my prior explicit written permission. This 
email and any files transmitted with it are confidential and intended 
solely for the use of the individual or entity to whom they are 
addressed. If you have received this email in error please notify the 
system manager. Please note that any views or opinions presented in this 
email are solely those of the author and do not necessarily represent 
those of the company. Finally, the recipient should check this email and 
any attachments for the presence of viruses. The company accepts no 
liability for any damage caused by any virus transmitted by this email.


More information about the volt-nuts mailing list