Hello Ken,

Which board are you looking at? On the amplifier board (A7) C24 is the 
input cap. Where is C8?
Don't rely on the multimeter to filter out the AC on the DC ranges, it's 
best to use a CRO. How is the signal coming into the amplifier from the 
oscillator? If the input to the amplifier is disconnected, is the offset 
still there?

Regards
On 10/07/14 20:42, Kgoodhew wrote:
> Hi Gerd and Charles,
> 	Thanks for the input it has given me other things to think about.
> I checked the output of pin 6 of U1 and I get a voltage variation of 10.2 mv
> from +15 to -15v adjustment on R6, so that appears to be working correctly.
> Now the fun starts :- using my HP 3478A to measure the dc on the output
> terminals I get  a reading of 0.9mv on the 1 volt range, 0.3 mv on the 10
> volt range, and 1.8 mv on the 100volt range.
> As a quick check of wether C8 may be leaky I pulled Q38 out of it's socket
> and got a constant DC out on the output terminals of 0.22mv on all three
> ranges, so maybe C8 is leaky?
> Using a fluke 77 3.5 digit multimeter yields vastly different DC readings,
> up to 5.5 v on the 100 volt range, so I suspect the Fluke 77 is reading some
> of the AC component on the output terminals even though DC is selected.
> Last night I did check the input and output on the amplifier, and on the 1 v
> range I was getting an input of 1v and an output of 3 volts, so the gain was
> correct, I did not check it on the 100v range though to see if the gain was
> 30.
> The input and output was a good clean sine wave on the amplifier.
> I have previously checked the 190 volt + and - supplies and they were spot
> on, but I will check them again before I pull the amplifier board apart
> again to remove C1 and C8 to test with a higher voltage other than the
> battery voltage of the multimeter, to see if they may be leaking at a higher
> voltage.
> Thanks,
> Ken
>
> -----
>
> Message: 2
> Date: Wed, 09 Jul 2014 14:21:13 -0400
> From: Charles Steinmetz <csteinmetz at yandex.com>
> To: Discussion of precise voltage measurement <volt-nuts at febo.com>
> Subject: Re: [volt-nuts] Fluke 5200A repair
> Message-ID: <20140709222123.LMBGV5Je at smtp2o.mail.yandex.net>
> Content-Type: text/plain; charset="us-ascii"; format=flowed
>
> Ken wrote:
>
>> R6 slider to U1 varies from +15v to -15v but the output of U1 only
>> varies by mv's, I put this down to the loading of R9 to earth on the
>> input to pin3 of U1 (221 ohm),only allowing a small variation to the
>> input on pin
>> 3 of U1, maybe this needs further investigation!
> Look at the circuit.  The 100k potentiometer (R6) feeds a voltage divider
> composed of R8 and R9 (464k and 221 ohms, respectively).  The tap of this
> divider feeds the reference input of U1.  R8 and R9 attenuate the +/- 15v
> range of R6 by a factor of 210, so the divider tap voltage (i.e., the
> reference voltage for U1) only varies +/- 7.14 mV from ground for the full
> range of R6.
>
> U1 is an integrator, so it has lots of gain at DC (Q38 switches the
> integrating capacitor to give two time constants).  If one of these
> capacitors is leaky or shorted (or there is another leakage path from Pin 6
> to Pin 2 of U1), it would reduce the gain of the integrator and could
> produce the symptom you are having.
>
> Best regards,
>
> Charles
>
>
>
>
>
> ------------------------------
>
> Message: 3
> Date: Wed, 09 Jul 2014 14:29:40 -0400
> From: Charles Steinmetz <csteinmetz at yandex.com>
> To: Discussion of precise voltage measurement <volt-nuts at febo.com>
> Subject: Re: [volt-nuts] Fluke 5200A repair
> Message-ID: <20140709222949.TnBeU5B8 at smtp2o.mail.yandex.net>
> Content-Type: text/plain; charset="us-ascii"; format=flowed
>
> Oops!  the divider ratio of R8/R9 is 2100, not 210.
>
> Best regards,
>
> Charles
>
>
>
> Message: 4
> Date: Thu, 10 Jul 2014 18:25:32 +1000
> From: Gerd <admin at controlelectronics.com.au>
> To: Discussion of precise voltage measurement <volt-nuts at febo.com>
> Subject: [volt-nuts]  Fluke 5200A repair
> Message-ID: <53BE4DFC.7090902 at controlelectronics.com.au>
> Content-Type: text/plain; charset=ISO-8859-1; format=flowed
>
> Hello Ken,
>
> The offset adjustment is about +-7mV to the input of U1. The amplifier has a
> gain of 3 or 30 so adjustment is small. Is the problem the same on both the
> 10V and 100V ranges? Is the amplifier producing the full output voltage? Is
> it distortion free?
> Are both the 190V regulators working correctly?
>
> Regards
>
>
>
> ---
> This email is free from viruses and malware because avast! Antivirus protection is active.
> http://www.avast.com
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
>

-- 
sig.html *Control Electronics Pty Ltd*
The home of Shiftmaster
Quality used test equipment
*Phone*: 02 9456 0736
*Website*: www.controlelectronics.com.au 
<http://www.controlelectronics.com.au/>*
Email*: admin at controlelectronics.com.au 
<mailto:admin at controlelectronics.com.au>
ABN 86 109 858 022
------------------------------------------------------------------------
The email addresses in this message are not to be added to any database, 
list, or mailing list without my prior explicit written permission. This 
email and any files transmitted with it are confidential and intended 
solely for the use of the individual or entity to whom they are 
addressed. If you have received this email in error please notify the 
system manager. Please note that any views or opinions presented in this 
email are solely those of the author and do not necessarily represent 
those of the company. Finally, the recipient should check this email and 
any attachments for the presence of viruses. The company accepts no 
liability for any damage caused by any virus transmitted by this email.


More information about the volt-nuts mailing list