[volt-nuts] Test Leads

acbern at gmx.de acbern at gmx.de
Tue Jul 21 16:19:15 EDT 2015


the best is really to use bare copper wires, such as used for wiring bells at home. does not look fancy, but low EMF.
Re. the low EMF Pomona cables with 4mm plugs, when you insert them in their (Pomona's) low EMF binding terminals, the spring that activates the safty protection pushes them out slowly, same with other binding posts, so not a safe connection if you are not carefull. Badly engineered imo.
I made my own cables with spades, using high quality gold plated pure copper spades and kapton twisted shielded cables, but again, bare copper is better (see guildline video on youtube re. their resistance bridges), but you will probably, if at all, only see that with a 34420A in lowest resolution (actually I have not).



> Gesendet: Dienstag, 21. Juli 2015 um 18:33 Uhr
> Von: "Randy Evans" <randyevans2688 at gmail.com>
> An: "volt-nuts at febo.com" <volt-nuts at febo.com>
> Betreff: [volt-nuts] Test Leads
>
> I recently acquired a Fluke 752A to go with my Fluke 732A and HP-3458A.
> Now I need to address the need for low thermal test leads.  Does anyone
> have any suggestions for test leads that have low thermal contributions to
> measurements?  I have looked at Pomona 1756 Low Thermal EMF Cables as a
> starting point but haven't found any other candidates.
> 
> I have heard that old TV twin lead works well since most are stranded
> copper wire.  I have tried it between my 732A and 3458A and it seems to
> work fine but I would like to use more professional looking test leads,
> particularly with correct copper spade lugs (Pomona 2305 Low thermal EMF
> spade lug, Gold-plated?).  Any other suggestions?
> 
> Thanks,
> 
> Randy Evans AE6YG