[volt-nuts] Fluke 720a self calibration repair

Bill Gold wpgold3637 at att.net
Tue Sep 29 03:08:34 EDT 2015


David:

    I just looked inside my 720A at the 20 turn 5K trimpots.  They are
Bourns 3005P-502 printed on them, cermets, standard 3/4", tempco 50 ppm.  So
the replacement is available directly from stock and nothing fancy.  Mine
are date code 8412 so my unit was built somewhere in the mid '80s.  Not too
sure where the idea of a wirewound pot came from here.  They are more than a
little pricey at $13.98 ea from Allied.  Looks like you need 24 to replace
them all in the "A" and "B" decades.  So that would be $335.52 unless you
buy 25 or more at which time the price drops to $12.51 which totals $312.75.
They come in tubes of 25 each which accounts for the price drop.

    Digikey is $14.35 and Mouser is $14.35.  Wow!  Expensive little devils.
On the other hand (left) Digikey shows a unit from "TT Electronics/BI
Technologies" for $1.40 ea.  Interesting.  Why such a massive difference in
price?  These are 100 ppm but as I have noted in another posting the tempco
isn't very important.  It's the resistors in the oil filled can that set the
overall stability.

Bill

----- Original Message ----- 
From: "David Garrido" <d.garrido at me.com>
To: <volt-nuts at febo.com>
Sent: Saturday, September 26, 2015 5:05 AM
Subject: Re: [volt-nuts] Fluke 720a self calibration repair


> Thank you for all of the input on this folks.
>
> Does anyone have a copy of the datasheet for the original Ramo-Eltra P/N
3800P-502?
>
> The reason I ask is that the price of these parts is driven in part by the
tempco and I found info that indicated 20ppm/C on one website, but that was
not a DS, just a listing of the info.  If they can be 50ppm/C or even
100ppm/C, then the price drops exponentially.
>
> I will do whatever is necessary to maintain or (LOL) improve the 720’s
legendary performance.
>
> Cheers,
>
> David