[volt-nuts] Update on 720A
Chuck Harris
cfharris at erols.com
Tue Aug 8 08:33:48 EDT 2017
Hi Dave,
Fundamentally, any input impedance from your 3458 is going
to throw your 720A out of wack. the 720A is not supposed to
have even picoamps of loading.... nothing, nada, zip.
I would suggest that before you damage your 720A further,
you read the manual for the 720A and follow the instructions
to the letter. An HP3458A is not a substitute for a null
detector and a voltage calibrator.
The calibration of a 720A requires that no current, none, zip,
be drawn from the 720A on any port. It can only be done by
using a voltage calibrator, and a null detector, as a
differential voltmeter.
Remember, and heed my words, *no* current may be drawn from the
720A.
-Chuck Harris
David C. Partridge wrote:
> Update: After changing the resistor I added in series with R1008 (A decade
> position 1.0) from 204.8 to 202.4 Ohms (I wonder if this means that R302 is slowly
> increasing in value back towards nominal) and re-calibrating again, the A decade
> is now within 1ppm of linear (using the 10V range and nominally 10V input).
>
> I note that when I'd done this, with the A decade set to 0.9 (reading 9.000000V on
> the 3458A), the actual input voltage needed was 10.000040 or so. Is that to be
> expected?
>
> Two of the positions on the B decade wouldn't quite adjust for a null when
> calibrating it. In this situation I think I have two options:
>
> 1) Add series resistors to compensate for slightly low value 9.898k resistors in
> the relevant two positions of the decade and recalibrate.
>
> 2) Adjust R203 to change the bridge balance slightly to the 10K ohm decades and
> recalibrate.
>
> Your thoughts on which approach to take is much appreciated.
>
> Thanks Dave
>
> _______________________________________________ volt-nuts mailing list --
> volt-nuts at febo.com To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the
> instructions there.
>
More information about the volt-nuts
mailing list