[volt-nuts] 3420B chopper drive problem
Dr. Frank
frank.stellmach at freenet.de
Tue Nov 7 12:14:35 EST 2017
Another probable error in the manual, on page 5-8 (80), diagram, which
shows the monitoring of the neon currents:
That diagram is ok for the A-version, as both neon currents are measured
with the same polarity in its oscillator circuit.
The B-version anyhow should look different, as the currents should be of
different polarity, so a positive pulse with about +150mV, and a
negative pulse with -75mV (or vice versa, depending on the polarity of
the diodes), always a signal pause at 0V.
One neon current is monitored over 51 Ohm, the other one over two 51Ohm
reistors in series, therefore these different amplitudes.
Frank
More information about the volt-nuts