[volt-nuts] 3458A question

Poul-Henning Kamp phk at phk.freebsd.dk
Wed Jan 10 05:23:50 EST 2018


--------
In message <[email protected]>, "David C. Partridge
" writes:

>After running an ACAL 0 with a four wire short in place I get a DCV reading
>starting at about  -0.00005 mV to increasing to about 0.00060 to 0.00070 mV
>over a minute or so at which point it seems to stabilise.

That is pretty normal, it's very hard to control temperatures so precisely
that you don't get a few hundred nanovolts of seebeck voltage.

If you touch the short with your finger, you will change the temperature
enough that it registers clearly.

I mount the short, then hang a piece of cloth over the terminals and wait
about 15-30 minutes before I run ACAL 0



-- 
Poul-Henning Kamp       | UNIX since Zilog Zeus 3.20
phk at FreeBSD.ORG         | TCP/IP since RFC 956
FreeBSD committer       | BSD since 4.3-tahoe    
Never attribute to malice what can adequately be explained by incompetence.