[volt-nuts] Prema 6048 zero offset problems
Bruce Griffiths
bruce.griffiths at xtra.co.nz
Thu Jan 18 07:54:46 EST 2018
Result of nonlinearity and dc offset combined?
Bruce
>
> On 19 January 2018 at 00:57 "David C. Partridge" <david.partridge at perdrix.co.uk> wrote:
>
> Starting with a warmed up Prema 6048, if I power it off and then on, select
> 20V DC range, 4 second integration and connect to the 10V output of a Fluke
> 732A, I get a reading of:
>
> +10.000047
>
> If I reverse the polarity of the cables and allow the reading to settle down
> for a minute, I get a reading of:
>
> -10.000006
>
> If I short the inputs of the meter, I get a reading of:
>
> +0.000105
>
> which is different from the +0.000020 which I would have expected from the
> readings of the 732A output.
>
> Proceeding on the assumption that the meter has stored a zero offset of
> -105uV
>
> I then press the ZERO button to null out the offset and then reconnect the
> 732A in normal polarity I get an incorrect reading of:
>
> +10.000153
>
> This is not what I would expect to see - I would expect to see a reading of
> +10.000047 - 0.000105 = +09.999942 This calculation
>
> I then reverse the inputs and get a reading of:
>
> -10.000113
>
> which is close to the expected reading of -10.000006 - 0.000105 = -10.000111
>
> I have mailed Prema about this problem, and will let you know what they have
> to say.
>
> Cheers
> David Partridge
>