> Make the adjustment in the correct sequence:
>
> 1. Set the Cal jumper to Cal Enable
> 2. Connect the reference to Ext Ref (rear) and Input A
> 3. Switch the counter to Ext Ref
> 4. Set the CalByte 50 for the best display (this is a very fine adjustement)
> 5. Switch the counter to Int Ref
> 6. Set the CalByte 4 for the best display (this adjustement is coarse,
> optimize!)
> 7. Set the Cal jumper to Cal Disable
>
> After this calibration my SR620 is
>
> with the external reference +/- 3 counts (9'999'999.9997x....10'000'000,0003)
> previous value -210 counts
> with the internal reference +/- 6 counts
> previous value -25 counts